[Intl_DxMedPhys] TJC acquisition monitor requirements
Januseski, Daniel
DJANUSESKI at virtua.org
Wed Sep 23 10:37:23 EDT 2026
I have been performing a test for years on CT where I measure a known object across the phantom with the on device distance tool and compare it to the real size of the object. For the ACR phantom, the included 100 mm point objects in section 3 (uniformity) seems optimum for this test. It is testing if the device’s display measuring tool is properly calibrated (footnote: never had a failure!).
Yes, technically this is a software feature and is not purely display-dependent so maybe this is not what they are considering. But considering display geometric accuracy as described by Bob is as he says dependent on aspect ratio and is not a factor in modern flat panel displays (it was a concern in CRTs due to EMF interference), this is my best guess at what JC wants.
Dan Januseski, MS, DABR
Voorhees, NJ
From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu> On Behalf Of Bob Kobistek via Intl_dxmedphys_wd_osu_list
Sent: Wednesday, September 23, 2026 9:32 AM
To: Douglas Pfeiffer <xraydoug at me.com>; DxMedPhys List <intl_dxmedphys_wd_osu_list at lists.osu.edu>
Subject: [EXTERNAL] Re: [Intl_DxMedPhys] TJC acquisition monitor requirements
**WARNING** This email did NOT originate from inside Virtua.
Please STOP and ANALYZE before opening attachments, clicking on links, or providing any information.
I’d forgotten about that. A while back I used to check it by measuring the length of some number of squares on a SMPTE pattern in both vertically and horizontally. So it was more of an aspect ratio test, but that’s the only thing I could think
ZjQcmQRYFpfptBannerStart
This Message Is From an External Sender
This message came from outside your organization.
Report Suspicious <https://us-phishalarm-ewt.proofpoint.com/EWT/v1/A_-8kG5y!DsgIW9oRIIIlTvUzwA4OJuEGPPg3XcsNVtRADuGjS9icmeTz1YRC7U2HwesDmt5HyRT2AuWut8H-8KLcMwhuTft2PlFTBk6CXM4rH7FfW2CCcLS0kEUVViutA6cnVHlR6jQ3fHerWwJHml01Ig$ >
ZjQcmQRYFpfptBannerEnd
I’d forgotten about that. A while back I used to check it by measuring the length of some number of squares on a SMPTE pattern in both vertically and horizontally. So it was more of an aspect ratio test, but that’s the only thing I could think
I’d forgotten about that. A while back I used to check it by measuring the length of some number of squares on a SMPTE pattern in both vertically and horizontally. So it was more of an aspect ratio test, but that’s the only thing I could think of that could be called spatial accuracy as it relates only to the monitor.
Robert J. Kobistek, MS, FACR, DABR, MRSE(MRSC™)
Medical Physicist
RJK Medical Physics, Inc.
440-463-7879
From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu>> On Behalf Of Douglas Pfeiffer via Intl_dxmedphys_wd_osu_list
Sent: Wednesday, September 23, 2026 8:31 AM
To: DxMedPhys List <intl_dxmedphys_wd_osu_list at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>>
Subject: [Intl_DxMedPhys] TJC acquisition monitor requirements
Good morning! I just received a document to prepare us for upcoming TJC inspections. It includes the following: My question is about the last item - spatial accuracy. I’ve never measured spatial accuracy on an acquisition display, and I don’t
Good morning!
I just received a document to prepare us for upcoming TJC inspections. It includes the following:
[cid:image001.png at 01DD4B47.83EF6B50]
My question is about the last item - spatial accuracy. I’ve never measured spatial accuracy on an acquisition display, and I don’t remember anyone ever talking about it. The only time I could imagine it being an issue at all is during MR- and CT-guided biopsies, but that is accuracy of the system calipers and not a function of the display. What are folks doing with this one?
Best regards,
Doug
Douglas Pfeiffer, MS, DABR FACR, FAAPM (he/him)
Medical Physicist, Radiation Safety Officer
Boulder Community Health
xraydoug at me.com<mailto:xraydoug at me.com>
303.415.7515
+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-
The opinions expressed in this message are the product of the gray and white matter loitering in my cranium. I speak for myself and no one else, unless I say otherwise.
This message, and any included attachments, are from Virtua Health or its related affiliates and is intended only for the addressee(s). The information contained herein is privileged, proprietary or may include confidential information and/or protected patient health information. Any unauthorized review, forwarding, printing, copying, distributing, or otherwise disseminating or taking any action based on such information is strictly prohibited. If you have received this message in error, or have reason to believe you are not authorized to receive it, please delete this message promptly and notify the sender by e-mail with a copy to ISSECURITY at virtua.org.
Thank you
This message was secured by Zix(R).
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://lists.osu.edu/pipermail/intl_dxmedphys_wd_osu_list/attachments/20260923/43740b22/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image001.png
Type: image/png
Size: 39407 bytes
Desc: image001.png
URL: <http://lists.osu.edu/pipermail/intl_dxmedphys_wd_osu_list/attachments/20260923/43740b22/attachment.png>
More information about the Intl_dxmedphys_wd_osu_list
mailing list