[Intl_DxMedPhys] ACR PET Count Rate Performance Test
Shane McCotter
shane at rpcphysics.com
Tue Aug 4 16:26:24 EDT 2026
Thank you to all who replied and helped me with the PET Count Rate Performance Test.
To follow up from my previous inquiry - I needed to use a 5.64 kg patient with the boost dose rather than using the same 70 kg patient from the initial phantom patient. After correcting for assay time and residual, I am getting within 10% of original.
A few of you reached out regarding what I found, if anyone wants my spreadsheet for calculating the count rate performance (and other PET tests), let me know!
Thanks!
Shane
________________________________
From: Shane McCotter
Sent: Tuesday, June 23, 2026 3:32 PM
To: DxMedPhys List <intl_dxmedphys_wd_osu_list at lists.osu.edu>
Subject: ACR PET Count Rate Performance Test
All,
Wondering if someone could walk me through the PET Count Rate Performance test as outlined in the ACR PET QC Manual. I have had trouble getting it to pass on a Siemens unit and am wondering if it is my scanner settings (dose, injection time, etc.).
Feel free to email privately if that is more desired.
Thanks,
Shane