[Intl_DxMedPhys] ACR MRI Philips Prescribed Phase FOV Question

Meier, Joseph Joseph.Meier at cchmc.org
Thu Jul 2 17:24:44 EDT 2026


Hello,

For ACR MRI submission, our Philips body/spine images have a prescribed FoV in the phase direction that is larger than the reconstructed phase FoV. (e.g., percent phase FoV is 175) to prevent aliasing.

Since the ACR form calculates the in plane resolution, I assume we should enter in the Reconstructed Phase and frequency FoVs for our submission as follows:



Freq FoV = 280

Prescribed Phase FoV = 490

Percent Phase FoV = 175

Phase FoV to be entered on the ACR data sheet should be 280 e.g., 490/1.75 = 280

Thank you for your insight,

Joe Meier



[cid:bfde3bb7-9a0f-493a-8557-36370e979f05]


Joseph Meier, PhD

Diagnostic Imaging Medical Physicist, Cincinnati Children’s Hospital

Assistant Professor, University of Cincinnati

3333 Burnet Ave, MLC 5031

Cincinnati, OH 45229-3026

(713) 422-3597 (O)

(513) 636-7775 (Administrative Assistant)





[cid:489d49e8-b19d-436d-bd62-c2413508b4bc]




-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://lists.osu.edu/pipermail/intl_dxmedphys_wd_osu_list/attachments/20260702/4d8436dd/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image.png
Type: image/png
Size: 53706 bytes
Desc: image.png
URL: <http://lists.osu.edu/pipermail/intl_dxmedphys_wd_osu_list/attachments/20260702/4d8436dd/attachment.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: Outlook-lzyzza4m.png
Type: image/png
Size: 29482 bytes
Desc: Outlook-lzyzza4m.png
URL: <http://lists.osu.edu/pipermail/intl_dxmedphys_wd_osu_list/attachments/20260702/4d8436dd/attachment-0001.png>


More information about the Intl_dxmedphys_wd_osu_list mailing list