When I started testing this camera in 2019, there already was an ACR folder under the USER folder, which had a protocol in it for testing the small phantom. I don’t know if it came from the manufacturer, or if my predecessor had made it by customizing a clinical protocol and saving it.
That being said, here are my notes on doing the postprocessing on the Xeleris station (which is the hard part), once you do get it to run and finish.
You will hopefully forgive if the notes are incomplete, or if I skipped parts. I wrote these for myself so I don’t forget. But at least you have a starting point.
@Maynard: I sent you a slightly longer version file. The listserv wouldn’t let me send it. Too big)
Nima
[cid:image001.png at 01DCA29E.11B7E870]
Nima Kasraie, PhD, MSc, DABR, DABSNM
Lead Physicist, Children’s Medical Center Dallas<https://urldefense.com/v3/__https://www.childrens.com/__;!!KGKeukY!2ItoqnIN4JwC3NWGgvXt13ifk2BOc04rorhIjouhZHxFKql0hyat-USv8ukEdL3Uc2anKK_LVK30qFgeSuNzNFlu4oWrrBhWoWppVlJzycny7JOoEMI2ng$ >
Associate Professor, Department of Radiology, UT Southwestern<https://urldefense.com/v3/__https://profiles.utsouthwestern.edu/profile/187443/nima-kasraie.html__;!!KGKeukY!2ItoqnIN4JwC3NWGgvXt13ifk2BOc04rorhIjouhZHxFKql0hyat-USv8ukEdL3Uc2anKK_LVK30qFgeSuNzNFlu4oWrrBhWoWppVlJzycny7JNJwGpyWg$ >
Phone: 214-648-7978
[cid:image002.png at 01DCA29E.11B7E870]
From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu> On Behalf Of High, Maynard via Intl_dxmedphys_wd_osu_list
Sent: Friday, February 20, 2026 10:22 AM
To: kenharper14 via Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu>
Cc: Erwin Ruff <erwin at marsdenmedphys.com>
Subject: [Intl_DxMedPhys] GE Discovery NM 530C Small Phantom SPECT Acquisition
We need any helpful tips on acquiring images of the SMALL Jayzak phantom for the first time as needed for ACR Accreditation SPECT annual physics report and image submission. This GE NM 530C is the small FOV digital camera that does not rotate
We need any helpful tips on acquiring images of the SMALL Jayzak phantom for the first time as needed for ACR Accreditation SPECT annual physics report and image submission. This GE NM 530C is the small FOV digital camera that does not rotate
We need any helpful tips on acquiring images of the SMALL Jayzak phantom for the first time as needed for ACR Accreditation SPECT annual physics report and image submission.
This GE NM 530C is the small FOV digital camera that does not rotate during acquisition.
Is it done with just a static exposure?
Which protocol is used to acquire and which application is used to process?
Any other tips would be appreciated.
Maynard High
Medical Physics: Westchester Medical Center - NYMC
914-493-1410
maynard.high at wmchealth.org<mailto:maynard.high at wmchealth.org>
________________________________
UT Southwestern
Medical Center
The future of medicine, today.
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20260221/1e641f46/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image001.png
Type: image/png
Size: 10985 bytes
Desc: image001.png
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20260221/1e641f46/attachment.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image002.png
Type: image/png
Size: 31503 bytes
Desc: image002.png
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20260221/1e641f46/attachment-0001.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: Instructions - Testing Summary GE530 small ACR phantom (clean).pdf
Type: application/pdf
Size: 2406524 bytes
Desc: Instructions - Testing Summary GE530 small ACR phantom (clean).pdf
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20260221/1e641f46/attachment.pdf>
More information about the Intl_dxmedphys_wd_osu_list
mailing list