[Intl_DxMedPhys] EXT MSG ACR CT Pediatric Clinical Image Question

Gauntt, David davidgauntt at uabmc.edu
Tue Apr 1 16:35:59 EDT 2025


The ACR has a pass/fail limit of 10mGy (32cm phantom) that applies  to 40-50lb patients, but there is no limit listed for clinical exams on larger patients (Technique Parameters and Anatomic Coverage: CT - Pediatric Abdomen Module (Revised 12-21-2021) : Accreditation Support<https://urldefense.com/v3/__https://accreditationsupport.acr.org/support/solutions/articles/11000049642-technique-parameters-and-anatomic-coverage-ct-pediatric-abdomen-module-revised-12-21-2021-__;!!KGKeukY!zv9cCpr-T2jYtsr4tMBua_SsApQGWgx-pcG0ETxyi0UHpdCGnrD6tsWAw1W-M8AgWjQ1t7Lp4iO3WtjBm_gLU1_OU7F-6oHFsssBKBCzrOTjog$ >).



However, the same page states “The facility will not receive a pass if the radiation dose is perceived to be too high”  (emphasis is mine).



Phantom testing is more explicit (Radiation Dosimetry: CT (Revised 3-25-2024) : Accreditation Support<https://urldefense.com/v3/__https://accreditationsupport.acr.org/support/solutions/articles/11000056198-radiation-dosimetry-ct-revised-3-25-2024-__;!!KGKeukY!zv9cCpr-T2jYtsr4tMBua_SsApQGWgx-pcG0ETxyi0UHpdCGnrD6tsWAw1W-M8AgWjQ1t7Lp4iO3WtjBm_gLU1_OU7F-6oHFsssBKBDz8yoQXw$ > ).  The tube current should be what you would use on a 40-50lb patient (and by extension the rotation time and helical pitch) in the site scanning form, which must match both the IQ and the dosimetry acquisitions.  The pass/fail limit is again 10mGy (32cm).


---
David M. Gauntt, Ph.D.
Associate Professor
Division of Physics and Engineering
Department of Radiology
(205)975-3777

[1647458080270]



________________________________
From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu> on behalf of Johnson, Loretta M via Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu>
Sent: Tuesday, April 1, 2025 2:31 PM
To: Timmerman,Michael; intl_dxmedphys_wd_osu_list at lists.osu.edu
Subject: Re: [Intl_DxMedPhys] EXT MSG ACR CT Pediatric Clinical Image Question

I don’t think there’s a pass/fail criterion, but you can look in the in the Radiology 2021 article on diagnostic reference levels by Kanal, et al. Here are the larger groups they have: 5 to < 10 years old 10 to < 15 15 to 18 Also available
I don’t think there’s a pass/fail criterion, but you can look in the in the Radiology 2021 article on diagnostic reference levels by Kanal, et al.  Here are the larger groups they have:

5 to < 10 years old
10 to < 15
15 to 18
Also available by effective diameter in a variety of pediatric sizes.

Peace,
Loretta

Loretta Johnson, Ph.D., DABR, FAAPM | Assistant Professor
Director | Division of Physics and Engineering
Department of Radiology | Heersink School of Medicine
UAB | The University of Alabama at Birmingham
GSB 301 | 619 19th Street South | Birmingham, AL 35249
P: 205.934.7592 | lorettajohnson at uabmc.edu<mailto:lorettajohnson at uabmc.edu>
Pronouns: she, her, hers

From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces+lorettajohnson=uabmc.edu at lists.osu.edu> on behalf of "Timmerman, Michael via Intl_dxmedphys_wd_osu_list" <intl_dxmedphys_wd_osu_list at lists.osu.edu>
Reply-To: "Timmerman,Michael" <Michael.Timmerman at SanfordHealth.org>
Date: Tuesday, April 1, 2025 at 2:07 PM
To: "intl_dxmedphys_wd_osu_list at lists.osu.edu" <intl_dxmedphys_wd_osu_list at lists.osu.edu>
Subject: EXT MSG [Intl_DxMedPhys] ACR CT Pediatric Clinical Image Question

Hi all, It’s been a while since I’ve reviewed clinical images for ACR accreditation. I have a quick question regarding pediatric abdomen pelvis exams. The pass/fail criteria is 10 mGy for the 32 cm phantom but this value is for a
Hi all,

It’s been a while since I’ve reviewed clinical images for ACR accreditation. I have a quick question regarding pediatric abdomen pelvis exams. The pass/fail criteria is 10 mGy for the 32 cm phantom but this value is for a 40-50 lb. patient; are there any guidelines for pass/fail criteria for older (and therefore larger) pediatric patients who may exceed this level?

Thanks!

Best,

Mike
________________________________________
Mike Timmerman, MS, DABR
Diagnostic Medical Physicist
Sanford Health – Sioux Falls Region
Mobile: (541) 213-1075
[Icon  Description automatically generated with medium confidence]


-----------------------------------------------------------------------
Confidentiality Notice: This e-mail message, including any attachments,
is for the sole use of the intended recipient(s) and may contain
privileged and confidential information. Any unauthorized review, use,
disclosure or distribution is prohibited. If you are not
the intended recipient, please contact the sender by reply e-mail and destroy
all copies of the original message.
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20250401/01d02b65/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: OutlookEmoji-1647458080270b6bd4a6f-a69b-4696-ae9e-d50df48a3e84.png
Type: image/png
Size: 15009 bytes
Desc: OutlookEmoji-1647458080270b6bd4a6f-a69b-4696-ae9e-d50df48a3e84.png
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20250401/01d02b65/attachment.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image001.png
Type: image/png
Size: 5514 bytes
Desc: image001.png
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20250401/01d02b65/attachment-0001.png>


More information about the Intl_dxmedphys_wd_osu_list mailing list