[Intl_DxMedPhys] Fw: Digirad SPECT - Jaszczak phantom for ACR
Joseph Giardina
jgiardina at petroneassoc.com
Tue Oct 20 11:02:41 EDT 2026
Subject: Re: Digirad SPECT - Jaszczak phantom for ACR
Anu
First thing is you actually want to center the phantom on the chair., not necessarily get as close to the detectors as possible
The other thing ive found is that the completed study defaults to an OSEM filter, so you should be able to reprocess it using a filtered FBP with the acr recommended butterworth filter which will make , especially the rods , look much better
Joe
Joseph Giardina, MS
Petrone Associates ( One Physics)
NYC
Good morning,
I an looking for some helpful tips and insights on acquiring ACR phantom images on the Digirad SPECT unit. I tested a Digirad X-act last week for the first time and followed the procedure that was provided by the company using the large phantom. The images were no where close to submission quality. Any suggestions would be greatly appreciated.
Thank you.
Best Regards,
Anu Soundararajan, PhD, DABR, DABSNM
Trinity Physics Consulting LLC