I also believe that the artifact on the spin echo ACR T1 is Gibbs truncation artifact, though it is quite pronounced.  I'm not very familiar with this scanner, but I would be inclined to disable filters (other than geometric distortion correction and image intensity uniformity correction).

The Site T1 FSE FLAIR sequence has a ring-like artifact that I often see in FSE imaging in the phase-encoding direction, more so on some scanners than others.  When I encounter this, I first disable flow compensation, if it is enabled, and then I try swapping frequency- and phase-encoding directions.  The width of and spacing between the rings should vary with echo train length and echo spacing - the tighter the echo train and the shorter the echo spacing, the narrower the rings.  Dropping phase matrix relative to the frequency matrix probably makes matters worse.  Admittedly, I am speculating a bit here...


_______________________________

John Greenhalgh, Ph.D.

Consulting Medical Physicist



E john.greenhalgh at westphysics.com<mailto:john.greenhalgh at westphysics.com>

T  866.275.WEST (9378)

O 770.435.9186  F 770.599.2333

W https://urldefense.com/v3/__http://www.westphysics.com__;!!KGKeukY!wjtw8htZhGVX8vYsMLCA_F9ycknNG08Ilv5B-zwYcB073m2eUXT7PjDDFpB-0Ohe4iAXPKtMKBdyuv1-n4gQkId2D7dPVcDKMY_sfwa_xt3evvsvWHYQ0Q$ <https://urldefense.com/v3/__http://www.westphysics.com__;!!KGKeukY!wjtw8htZhGVX8vYsMLCA_F9ycknNG08Ilv5B-zwYcB073m2eUXT7PjDDFpB-0Ohe4iAXPKtMKBdyuv1-n4gQkId2D7dPVcDKMY_sfwa_xt3evvsvWHYQ0Q$ >



[A picture containing drawing  Description automatically generated]<https://urldefense.com/v3/__http://www.westphysics.com/__;!!KGKeukY!wjtw8htZhGVX8vYsMLCA_F9ycknNG08Ilv5B-zwYcB073m2eUXT7PjDDFpB-0Ohe4iAXPKtMKBdyuv1-n4gQkId2D7dPVcDKMY_sfwa_xt3evvsVVj1QCA$ >



The information contained in this transmission may contain privileged and confidential information, including patient information protected by federal and state privacy laws. It is intended only for the use of the person(s) named above. If you are not the intended recipient, you are hereby notified that any review, dissemination, distribution, or duplication of this communication is strictly prohibited. If you are not the intended recipient, please contact the sender by reply email and destroy all copies of the original message.

________________________________
From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces+john.greenhalgh=westphysics.com at lists.osu.edu> on behalf of Tipnis, Sameer via Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu>
Sent: Wednesday, October 7, 2026 2:36 PM
To: Dave Jordan <dave.jordan at gmail.com>; High, Maynard <Maynard.High at wmchealth.org>; DxMedPhys List <intl_dxmedphys_wd_osu_list at lists.osu.edu> <intl_dxmedphys_wd_osu_list at lists.osu.edu>
Subject: Re: [Intl_DxMedPhys] MRI ACR Phantom Artifacts

Agree with the Gibbs artifact assessment for the ACR T1. It is possible that a high-pass filter was left on. We had this issue recently and had to turn it off before getting decent images. The Site T1 seems to be affected with motion. _________________________________________________

Agree with the Gibbs artifact assessment for the ACR T1. It is possible that a high-pass filter was left on. We had this issue recently and had to turn it off before getting decent images. The Site T1 seems to be affected with motion.



_________________________________________________



Sameer Tipnis, PhD, DABR, FAAPM

Professor of Radiology & Chief, Division of Medical Physics

Program Director, Diagnostic Medical Physics Residency

Dept. of Radiology and Radiological Sciences

Medical University of South Carolina

96 Jonathan Lucas Street, MSC 323

Charleston, SC 29425



Phone: (843)-792-3833

Fax:     (843)-792-2642

_________________________________________________





From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces+tipnis=musc.edu at lists.osu.edu> On Behalf Of Dave Jordan via Intl_dxmedphys_wd_osu_list
Sent: Tuesday, October 6, 2026 5:54 PM
To: High, Maynard <Maynard.High at wmchealth.org>; DxMedPhys List <intl_dxmedphys_wd_osu_list at lists.osu.edu> <intl_dxmedphys_wd_osu_list at lists.osu.edu>
Subject: Re: [Intl_DxMedPhys] MRI ACR Phantom Artifacts



Your ACR T1 has Gibbs artifact; normally you don't see that on those GEs, so there may be some low-pass filtering turned off in a service menu or control variable. Looks pretty normal for a Siemens or Philips to me.   Your Site T1 appears

Your ACR T1 has Gibbs artifact; normally you don't see that on those GEs, so there may be some low-pass filtering turned off in a service menu or control variable. Looks pretty normal for a Siemens or Philips to me.   Your Site T1 appears

Your ACR T1 has Gibbs artifact; normally you don't see that on those GEs, so there may be some low-pass filtering turned off in a service menu or control variable. Looks pretty normal for a Siemens or Philips to me.



Your Site T1 appears to have some vibration that is sensitive at that particular TR value. If you're not seeing it on your ACR sequences and your Site T2 that pretty much confirms it, but you can try running the Site T1 with a different TR (try, say, 1500 or 1900 ms) and confirm that it goes away.



---

On Oct 6, 2026 at 4:06 PM -0400, High, Maynard via Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>>, wrote:


MRI experienced Listers: I have come across some nasty artifacts in the ACR Phantom images for ACR accreditation images and wonder if anybody has any advice. The MRI is a 2014 GE Optima MR450W. All images were with PURE uniformity correction

MRI experienced Listers:

I have come across  some nasty artifacts in the ACR Phantom images for ACR accreditation images and wonder if anybody has any advice.



The MRI is a 2014 GE Optima MR450W.  All images were with PURE uniformity correction on and the coil was an HNS Head 12-Channel.

The phantom is the Medium sized.



First is the ACR T1 sequence:

What could be causing these rings?

<image.png>



Next is the Site FSE T1 Flair



<image.png>

Maynard High

Medical Physics: Westchester Medical Center - NYMC

914-493-1410

maynard.high at wmchealth.org<mailto:maynard.high at wmchealth.org>
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://lists.osu.edu/pipermail/intl_dxmedphys_wd_osu_list/attachments/20261008/45806899/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image.png
Type: image/png
Size: 107762 bytes
Desc: image.png
URL: <http://lists.osu.edu/pipermail/intl_dxmedphys_wd_osu_list/attachments/20261008/45806899/attachment.png>


More information about the Intl_dxmedphys_wd_osu_list mailing list