[Intl_DxMedPhys] ACR CT test question
Douglas Pfeiffer
xraydoug at me.com
Mon Sep 14 16:47:30 EDT 2026
The 21 cm DFOV is to make it easier to analyze - nothing to do with patient size. It doesn’t impact dose in any way. Using a small Scan FOV is to reduce non-uniformities due to the bowtie not being matched to the object size on some scanners.
Best regards,
Doug
Douglas Pfeiffer, MS, DABR FACR, FAAPM (he/him)
Medical Physicist, Radiation Safety Officer
Boulder Community Health
xraydoug at me.com
303.415.7515
+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-
The opinions expressed in this message are the product of the gray and white matter loitering in my cranium. I speak for myself and no one else, unless I say otherwise.
> On Sep 14, 2026, at 2:42 PM, Matt Wait <Matt.Wait at kp.org> wrote:
>
> That's good to hear, that's the practice I've been using. But I have never found a good reference for correlating the abdomen diameter (21 cm?) with 70 kg.
>
> Matt Wait, MS, DABR, DABSNM, MRSE
> Senior Diagnostic Physicist
> Associate Radiation Safety Officer
> Assistant Residency Director
> Kaiser Permanente
> Southern Permanente Medical Group
> Medical Imaging Technology and Informatics
> 4867 Sunset Blvd
> Los Angeles, CA 90027
> x5347 (office)
> (818) 232-2427 (mobile phone)
>
> From: Douglas Pfeiffer <xraydoug at me.com>
> Sent: Monday, September 14, 2026 1:35 PM
> To: Matt Wait <Matt.Wait at kp.org>
> Cc: DxMedPhys List <intl_dxmedphys_wd_osu_list at lists.osu.edu>; Bob Kobistek <bob at rjkmedphys.com>
> Subject: Re: [Intl_DxMedPhys] ACR CT test question
>
> Going through the list of scanned patients on the scanner and looking for abdomens that look about the right size, and looking at the mA at mid-liver (belly button-ish).
>
> Best regards,
> Doug
>
>
>
> Douglas Pfeiffer, MS, DABR FACR, FAAPM (he/him)
> Medical Physicist, Radiation Safety Officer
> Boulder Community Health
> xraydoug at me.com
> 303.415.7515
>
> +-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-
>
> The opinions expressed in this message are the product of the gray and white matter loitering in my cranium. I speak for myself and no one else, unless I say otherwise.
>
>> On Sep 14, 2026, at 2:16 PM, Matt Wait <Matt.Wait at kp.org> wrote:
>>
>> Doug, how did you envision determining the automatic tube current selection for Standard Man?
>>
>> Matt Wait, MS, DABR, DABSNM, MRSE
>> Senior Diagnostic Physicist
>> Associate Radiation Safety Officer
>> Assistant Residency Director
>> Kaiser Permanente
>> Southern Permanente Medical Group
>> Medical Imaging Technology and Informatics
>> 4867 Sunset Blvd
>> Los Angeles, CA 90027
>> x5347 (office)
>> (818) 232-2427 (mobile phone)
>>
>>
>> From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces+matt.wait=kp.org at lists.osu.edu <mailto:intl_dxmedphys_wd_osu_list-bounces+matt.wait=kp.org at lists.osu.edu>> on behalf of Douglas Pfeiffer via Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu <mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>>
>> Sent: Monday, September 14, 2026 1:12 PM
>> To: Bob Kobistek <bob at rjkmedphys.com <mailto:bob at rjkmedphys.com>>
>> Cc: DxMedPhys List <intl_dxmedphys_wd_osu_list at lists.osu.edu <mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>>
>> Subject: Re: [Intl_DxMedPhys] ACR CT test question
>>
>> Caution: This email came from outside Kaiser Permanente. Do not open attachments or click on links if you do not recognize the sender.
>> For what it’s worth now, the intent when I was working on the manual was Standard Man (i.e., 70 kg). It was an unfortunate oversight on my part that I did not catch that and state it explicitly way back when. I’m not sure what is assumed now by the ACR, but that was the original thinking.
>>
>> Best regards,
>> Doug
>>
>>
>>
>> Douglas Pfeiffer, MS, DABR FACR, FAAPM (he/him)
>> Medical Physicist, Radiation Safety Officer
>> Boulder Community Health
>> xraydoug at me.com
>> 303.415.7515
>>
>> +-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-
>>
>> The opinions expressed in this message are the product of the gray and white matter loitering in my cranium. I speak for myself and no one else, unless I say otherwise.
>>
>>> On Sep 9, 2026, at 1:45 PM, Bob Kobistek via Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu> wrote:
>>>
>>> Hi.
>>>
>>> I also beat my head against a wall searching for answers to these questions when the ACR CT accreditation program was started. I stopped overthinking it, and now I pretty much wing it.
>>>
>>> For average size patient, I look through patient scans to find one with an A-P dimension through the abdomen of around 23 cm. If I can find more than one (hard to find nowadays), I’ll toss out the outliers and take an average of the mAs values.
>>>
>>> For new scanners I use nearly all the options you presented. If it’s a scanner with a similar model in use at the facility, I refer to patients scanned on the existing scanners. If it’s a scanner that I see at other facilities, I’ll make adjustments for differences in protocols and use that. If it’s a scanner I’ve never seen before, I’ll pull up the manufacturer’s suggested protocols and take a guess at the mAs when TCM is used. If I use the second or third option, I’ll take a look at patient scans once the scanner has been operating for a few weeks. If I can make adjustments to my measurements based on the clinical data, I will. If I have to, I’ll repeat the measurements using the clinical protocols. If the site is going to be ACR accredited I will always make a second trip so I can use the actual protocols and clinical mAs values for accreditation, not matter which of the three options above that I use.
>>>
>>> There is very little guidance on how to chose mAs (and kVp when applicable) for automated technique selection. So we use out best judgment.
>>>
>>>
>>> Robert J. Kobistek, MS, FACR, DABR, MRSE(MRSC™)
>>> Medical Physicist
>>> RJK Medical Physics, Inc.
>>> 440-463-7879
>>>
>>> From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu <mailto:intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu>> On Behalf Of David S. Winter via Intl_dxmedphys_wd_osu_list
>>> Sent: Monday, September 7, 2026 4:02 PM
>>> To: DxMedPhys List <intl_dxmedphys_wd_osu_list at lists.osu.edu <mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>>
>>> Subject: [Intl_DxMedPhys] ACR CT test question
>>>
>>> All, I apologize as this has no doubt been discussed in the past, but I have a two part inquiry related to ACR CT testing. I am happy to clarify directly with the ACR, but I figured I would touch base with this forum first. (Especially since
>>> All,
>>>
>>> I apologize as this has no doubt been discussed in the past, but I have a two part inquiry related to ACR CT testing. I am happy to clarify directly with the ACR, but I figured I would touch base with this forum first. (Especially since only part 1 would be good to ask the ACR directly.)
>>>
>>> Part 1: The ACR’s scanning instructions webpage and the CT QC manual both give a weight range for the pediatric abdominal exam of 40-50lbs, but only the QC manual has a weight for the adult abdomen (i.e. 70 kg, ~154 lbs). The webpage merely says “average sized patient”. The exclusion of the 70 kg reference could be an accidental omission, but that requires the same omission to have been repeated in multiple different locations on their webpage.
>>>
>>> “Average sized” could be interpreted to mean average across the entire US population, the region a CT scanner is in, or the specific patient population a specific CT scanner sees (which could be massively different if the scanner is the go-to scanner for bariatric patients.) The CDC says the current adult average weights are 199 lbs for men and 171.8 lbs for women. (Both significantly more than 154). In addition, the ACR’s mandatory dose calculation sheet calculates the adult abdomen SSDE for a 35 cm water equivalent diameter. While you can’t exactly convert abdominal water equivalent diameter to a patient weight, that value would likely correspond to a patient significantly heavier than 154 lbs.
>>>
>>> Has anyone specifically seen anything that dictates which is definitively correct for determining what techniques to set, and by extension, determine whether or not the scanner is really passing the CNR test?
>>>
>>> Part 2: In practice, I am curious how people are determining the mA that CT scanners are selecting for the “average”, particularly for brand new scanners without much/any historic patient data. I can imagine four different general approaches. 1) Analysis of historical data aggregated by a dose metric monitoring system like Radimetrics. 2) Manually looking for a few historical patients that look to be roughly “average size” and then averaging the mA from a representative slice from a few of those studies. 3) For new systems, using values from a similar system with similar settings. 4) Conducting scans of a standard (e.g. their CTDI phantom), seeing what the CT scanner picks for mA, and then applying a scaling factor to that mA to adjust for the size difference.
>>>
>>> Sincerely,
>>> Scott
>>>
>>> D. SCOTT WINTER
>>>
>>> CHIEF DIAGNOSTIC MEDICAL PHYSICIST
>>> O:
>>> 228-865-3475 <tel:228-865-3475>
>>> 4500 13th St
>>> ,
>>> Gulfport
>>> ,
>>> MS
>>> ,
>>> 39501
>>> <image001.gif> <https://urldefense.com/v3/__https:/www.wearememorial.com/__;!!KGKeukY!2MSmyveKKzGn4EYDtlECE0uHMzo9qICIABIb_AilEMGb_zQ1uA7SIMU5B_TVMuEa8DaUCdbH70xPaPjVYos6Dt0dayhOFYF5vxZV$>
>>> WeAreMemorial.com <https://urldefense.com/v3/__https:/www.wearememorial.com/__;!!KGKeukY!2MSmyveKKzGn4EYDtlECE0uHMzo9qICIABIb_AilEMGb_zQ1uA7SIMU5B_TVMuEa8DaUCdbH70xPaPjVYos6Dt0dayhOFYF5vxZV$>
>>> |
>>> Facebook <https://urldefense.com/v3/__https:/www.facebook.com/MemorialHospitalMS/__;!!KGKeukY!2MSmyveKKzGn4EYDtlECE0uHMzo9qICIABIb_AilEMGb_zQ1uA7SIMU5B_TVMuEa8DaUCdbH70xPaPjVYos6Dt0dayhOFdPd1EsI$>
>>> |
>>> YouTube <https://urldefense.com/v3/__https:/www.youtube.com/user/gulfportmemorial__;!!KGKeukY!2MSmyveKKzGn4EYDtlECE0uHMzo9qICIABIb_AilEMGb_zQ1uA7SIMU5B_TVMuEa8DaUCdbH70xPaPjVYos6Dt0dayhOFUp5zZIl$>
>>> |
>>> LinkedIn <https://urldefense.com/v3/__https:/www.linkedin.com/company/memorial-hospital-at-gulfport__;!!KGKeukY!2MSmyveKKzGn4EYDtlECE0uHMzo9qICIABIb_AilEMGb_zQ1uA7SIMU5B_TVMuEa8DaUCdbH70xPaPjVYos6Dt0dayhOFbsk9WzQ$>
>>> This email and any files transmitted with it may contain PRIVILEGED or CONFIDENTIAL information and may be read or used only by the intended recipient. If you are not the intended recipient of the email or any of its attachments, please be advised that you have received this email in error and that any use, dissemination, distribution, forwarding, printing, or copying of this email or any attached files is strictly prohibited. If you have received this email in error, please immediately purge it and all attachments and notify the sender by reply email or contact the sender at the number listed.
>>
>> NOTICE TO RECIPIENT: If you are not the intended recipient of this e-mail, you are prohibited from sharing, copying, or otherwise using or disclosing its contents. If you have received this e-mail in error, please notify the sender immediately by reply e-mail and permanently delete this e-mail and any attachments without reading, forwarding or saving them. v.173.295 Thank you.
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://lists.osu.edu/pipermail/intl_dxmedphys_wd_osu_list/attachments/20260914/875707bb/attachment.html>
More information about the Intl_dxmedphys_wd_osu_list
mailing list