Steve,

I have not heard of this being an issue as well, but I imagine in CA the
tech could always perform the State required weekly QC after transport and
delivery for peace of mind.

For those unaware, a phantom, likely a metal plate, is imaged, and the kV
amd mA must be within a certain range, given by a dose ratio calculation
against a baseline (i.e. kV^2 * mA). For those who don't want to do a
clunky calculation, the units report dose rate. The idea is for the dose
rate to be within 25% of a baseline.

Steven

Steven M. LaFontaine, M.S., DABR
Certified Medical Physicist
Therapy Physics, Inc.
2501 Cherry Ave, Suite #270
Signal Hill, CA 90755
562-317-0650 (office)
310-720-5899 (cell)
Steven at therapyphysics.com


On Sat, Aug 22, 2026, 5:09 PM Gary via Intl_dxmedphys_wd_osu_list <
intl_dxmedphys_wd_osu_list at lists.osu.edu> wrote:

> Excellent point.  After all, it is portable.
> OTOH, all that packing moving and unpacking is bound to jiggle something.
> Might be smart for the RSO of that outfit to require some basic daily QC
> before clinical use.
>
>
> ---
> Thanks,
> Gary Isenhower, MS
>
>
> On Sat, 22 Aug 2026 19:22:46 +0000, Bob Kobistek via
> Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu>
> wrote:
>
> >One of my clients is a mobile lithotripsy company. They have 9 units
> >that go out every day in the Northeast Ohio area. The units are
> >brought into the client hospital’s surgery department, not left on the
> >truck. At the end of the workday, the units are returned to the
> >warehouse and readied for the next day. It would be totally
> >impractical for them to have a physicist follow each of those units
> >every day. We have to weigh the cost against the benefit.
> >
> >
> >Robert J. Kobistek, MS, FACR, DABR, MRSE(MRSC™)
> >Medical Physicist
> >RJK Medical Physics, Inc.
> >440-463-7879
> >
> >From: Intl_dxmedphys_wd_osu_list
> ><intl_dxmedphys_wd_osu_list-bounces+bob=rjkmedphys.com at lists.osu.edu>
> >On Behalf Of Nicole Ranger via Intl_dxmedphys_wd_osu_list Sent:
> >Saturday, August 22, 2026 10:44 AM To: Stephen Steuterman
> ><Steuterman.Stephen at scrippshealth.org> Cc: DxMedPhys List
> ><intl_dxmedphys_wd_osu_list at lists.osu.edu> Subject: Re:
> >[Intl_DxMedPhys] OEC 3D C-arm Testing
> >
> >My policy at a former facility in WI was that a portable device
> >transported within the facility only had annual testing but if that
> >device was transported offsite, it needed to be tested before use at
> >that new location. A device in a permanently My policy at a former
> >facility in WI was that a portable device transported within the
> >facility only had annual testing but if that device was transported
> >offsite, it needed to be tested before use at that new location.  A
> >device in a permanently fixed and secured position in a truck and
> >operated at a new location, would be different though arguably one
> >might want to perform more daily QC testing.
> >
> >Nicole
> >
> >
> >Sent from my iPhone
> >
> >
> >On Aug 21, 2026, at 1:23 PM, Steuterman, Stephen via
> >Intl_dxmedphys_wd_osu_list
> ><intl_dxmedphys_wd_osu_list at lists.osu.edu<mailto:
> intl_dxmedphys_wd_osu_list at lists.osu.edu>>
> >wrote:  Hello all, Portable C-arms are transported to multiple
> >locations frequently by lithotripsy companies. Typically, they have
> >their annual physicist testing, and I don't believe they do any
> >additional testing after each move. Are there different Hello all,
> >
> >
> >Portable C-arms are transported to multiple locations frequently by
> >lithotripsy companies.  Typically, they have their annual physicist
> >testing, and I don't believe they do any additional testing after each
> >move.   Are there different requirements for a 3D C-arm?  I have a
> >friend at an imaging center, who has a tech who wouldn't use a OEC 3D
> >C-arm after it was returned from another location.    The tech said
> >the FSE told her a EPE eval was required after transport and prior to
> >patient use.   I wasn't able to find that requirement in the user's
> >manual, and the FSE didn't give me the reference either.
> >
> >Do you do a full physicist test every time the 3D C-arm is transported?
> >
> >Thanks
> >
> >Steve
> >
> >
> >Stephen Steuterman, M.S.
> >
> >Director of Medical Physics & Radiation Safety
> >
> >System-Wide Radiation Safety Officer
> >
> >
> >
> >619-709-1546 (cell)
> >
> >
> >
> >Steuterman.Stephen at scrippshealth.org<mailto:
> Steuterman.Stephen at scrippshealth.org>
> >
> >
> >This e-mail and any files transmitted with it may contain privileged
> >and confidential information and are intended solely for the use of
> >the individual or entity to which they are addressed. If you are not
> >the intended recipient or the person responsible for delivering the
> >e-mail to the intended recipient, you are hereby notified that any
> >dissemination or copying of this e-mail or any of its attachment(s) is
> >strictly prohibited. If you have received this e-mail in error, please
> >immediately notify the sending individual or entity by e-mail and
> >permanently delete the original e-mail and attachment(s) from your
> >computer system. Thank you.
>
>
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://lists.osu.edu/pipermail/intl_dxmedphys_wd_osu_list/attachments/20260824/2366ea1b/attachment.html>


More information about the Intl_dxmedphys_wd_osu_list mailing list