[Intl_DxMedPhys] OEC 3D C-arm Testing

Bob Kobistek bob at rjkmedphys.com
Sat Aug 22 15:22:46 EDT 2026


One of my clients is a mobile lithotripsy company. They have 9 units that go out every day in the Northeast Ohio area. The units are brought into the client hospital’s surgery department, not left on the truck. At the end of the workday, the units are returned to the warehouse and readied for the next day. It would be totally impractical for them to have a physicist follow each of those units every day. We have to weigh the cost against the benefit.


Robert J. Kobistek, MS, FACR, DABR, MRSE(MRSC™)
Medical Physicist
RJK Medical Physics, Inc.
440-463-7879

From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces+bob=rjkmedphys.com at lists.osu.edu> On Behalf Of Nicole Ranger via Intl_dxmedphys_wd_osu_list
Sent: Saturday, August 22, 2026 10:44 AM
To: Stephen Steuterman <Steuterman.Stephen at scrippshealth.org>
Cc: DxMedPhys List <intl_dxmedphys_wd_osu_list at lists.osu.edu>
Subject: Re: [Intl_DxMedPhys] OEC 3D C-arm Testing

My policy at a former facility in WI was that a portable device transported within the facility only had annual testing but if that device was transported offsite, it needed to be tested before use at that new location. A device in a permanently
My policy at a former facility in WI was that a portable device transported within the facility only had annual testing but if that device was transported offsite, it needed to be tested before use at that new location.  A device in a permanently fixed and secured position in a truck and operated at a new location, would be different though arguably one might want to perform more daily QC testing.

Nicole


Sent from my iPhone


On Aug 21, 2026, at 1:23 PM, Steuterman, Stephen via Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>> wrote:

Hello all, Portable C-arms are transported to multiple locations frequently by lithotripsy companies. Typically, they have their annual physicist testing, and I don't believe they do any additional testing after each move. Are there different
Hello all,


Portable C-arms are transported to multiple locations frequently by lithotripsy companies.  Typically, they have their annual physicist testing, and I don't believe they do any additional testing after each move.   Are there different requirements for a 3D C-arm?  I have a friend at an imaging center, who has a tech who wouldn't use a OEC 3D C-arm after it was returned from another location.    The tech said the FSE told her a EPE eval was required after transport and prior to patient use.   I wasn't able to find that requirement in the user's manual, and the FSE didn't give me the reference either.

Do you do a full physicist test every time the 3D C-arm is transported?

Thanks

Steve


Stephen Steuterman, M.S.

Director of Medical Physics & Radiation Safety

System-Wide Radiation Safety Officer



619-709-1546 (cell)



Steuterman.Stephen at scrippshealth.org<mailto:Steuterman.Stephen at scrippshealth.org>


This e-mail and any files transmitted with it may contain privileged and confidential information and are intended solely for the use of the individual or entity to which they are addressed. If you are not the intended recipient or the person responsible for delivering the e-mail to the intended recipient, you are hereby notified that any dissemination or copying of this e-mail or any of its attachment(s) is strictly prohibited. If you have received this e-mail in error, please immediately notify the sending individual or entity by e-mail and permanently delete the original e-mail and attachment(s) from your computer system. Thank you.
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://lists.osu.edu/pipermail/intl_dxmedphys_wd_osu_list/attachments/20260822/f37d2edd/attachment.html>


More information about the Intl_dxmedphys_wd_osu_list mailing list