[Intl_DxMedPhys] MQSA - Manufacturer PM vs ACR DM QC manual
Tao, Ashley T
attao at emplifyhealth.org
Tue May 19 16:26:50 EDT 2026
Dear DxMedPhys Listerv,
I am wondering if anyone who is using the ACR DM QC manual has had any experience with MQSA inspectors giving citations for not following up with PM specifications.
Let me set the scene:
1.
Our biomed engineer performed service on our unit and determined HVL to be outside of Hologic specs of +/-0.01 mm Al from their baseline/reference value. At 28kV, hologic baseline was 0.513 mm Al, and biomed measured 0.517. We were due for annual testing about a month and a half from the discovery. biomed asked if they could schedule PM for the day before physics testing. Since we were well above the minimum HVL requirement of 0.28 mm Al, we agreed.
MQSA inspectors said that ACR DM QC Test summary (mee report), Beam quality (HVL) assessment failures must be corrected before clinical use. Our HVL passes the ACR DM QC manual, but because biomed failed it, they said we needed to correct before clinical use and have an MEE.
2.
Our biomed engineer performed service on our unit and determined AEC performance to be outside Hologic's specifications @ 2cm. Hologic references a range of EI values, whereas ACR DM QC manual specifies the SNR needs to be within +/-15% of the MEE. These two metrics are not comparable. We did perform a partial MEE because we were not certain whether or not we needed to. It turns out the MEE SNR was 22.34 and after AEC calibration, the new MEE SNR was 22.21. The MEE was completed 34 days after service but we were cited because it needed to be within 30 days. How do you reconcile the metrics used by Hologic vs ACR DM QC manual if they are not comparable?
3.
Our biomed engineer performed service on our unit and determined target dose to be outside Hologic's specifications of +/-0.1 mGy. The ACR DM QC manual specifies the target dose for a CC view of the ACR DM QC phantom to be <=3 mGy and that the unit AGD must be within +/-25% of the calculated AGD. Again, Hologics metrics are different than the ACR DM QC manual. We are well within specifications, but because biomed failed it and we did not service it within 30 days, we were cited.
We showed the MQSA inspectors the service report and explained what the results were but they didn't seem to care and said that we would need to provide a corrective action letter before escalating to their next level inspector. It is my understanding that Hologic PM specs are more stringent for optimal performance. The ACR DM QC manual does not specify a timeline to address deficiencies in PM except for if there is relevant service and a finding were to be out of compliance with the ACR DM QC manual then you would need appropriate follow-up (e.g. immediately before clinical use versus 30 days). We are writing a response letter addressing these citations, but I would appreciate any input this group may have regarding their experience w/ inspectors and PMs being out of spec.
On a side note, if HVL is out of Hologic spec (but within MQSA requirements), and my biomed engineer needs to recalibrate but it also requires them to recalibrate AEC as part of the process, it seems that there would need to be on-site physicist testing of the AEC because according to ACR DM QC manual - Table 3 Medical Physicist Involvement in Equipment Adjustments, Changes or Repairs(pg 138) - are any of you basically having to do MEEs every time your PMs are just a little bit out of spec?
Looking forward to hearing about your experiences.
Thanks
Ashley