[Intl_DxMedPhys] Discovery IQ PET/CT Protocol help

Nye, Jonathon nyej at musc.edu
Fri May 15 11:20:06 EDT 2026


Hi Mike,

Have you been able to verify the SUV inaccuracies with a phantom?

Or is this something that happens only in large patients?

In my experience, QClear and SharpIR can cause some overshoot of SUV in small lesions but that was something discovered when the algorithms were first introduced and may be different now.

Jon


Jonathon A. Nye, PhD

Professor, Department of Radiology and Radiological Science

Medical University of South Carolina
99 Jonathan Lucas Drive, MSC 184

Charleston, SC 29425
Phone: (843)-792-4368


________________________________
From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces+nyej=musc.edu at lists.osu.edu> on behalf of Mike Nimmo via Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu>
Sent: Thursday, May 14, 2026 10:58 AM
To: kenharper14 via Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu>
Subject: [Intl_DxMedPhys] Discovery IQ PET/CT Protocol help

Hello, I have a customer who has received delivery of a GE Discovery IQ PET/CT 5 ring system. Their PET radiologist is unhappy with the BGO system's PET image quality for medium to large patients. I was hoping someone out there with a 3

Hello, I have a customer who has received delivery of a GE Discovery IQ PET/CT 5 ring system. Their PET radiologist is unhappy with the BGO system's PET image quality for medium to large patients. I was hoping someone out there with a 3
Hello,

I have a customer who has received delivery of a GE Discovery IQ PET/CT 5 ring system. Their PET radiologist is unhappy with the BGO system's PET image quality for medium to large patients. I was hoping someone out there with a 3 or 5 ring PET/CT system could share their experience and maybe some protocol details with me to pass along.

Particular concerns involve SUV overshoot when detecting small lesions. Also the image quality on medium to large adults is noisy and unsatisfactory. Some things we've been trying to improve image quality are below:

1. Try turning on Q.Clear with Beta values of 200, 250, and 300
2. Increase bed overlap from 19 slices to 35 slices.
3. Changing bedstop acquisition time to 200 seconds or 400 seconds per bed stop for medium to large patients.
4. Using Z-axis standard for small to medium patients and Heavy for larger patients.
5. using Sharp IR

They are still struggling with image quality and SUV accuracy.

Additionally the radiologist appears doubtful whether this scanner will be effective for imaging PSMA or Dotatate studies.

I do not have much experience on this scanner platform so I would appreciate any assistance.

Thanks,

--
Mike Nimmo