[Intl_DxMedPhys] [Ext] Re: current MQSA policy on mammo RWS MEE testing

Wilson, Robert J rwilson at uthsc.edu
Wed Apr 1 11:29:30 EDT 2026


The small entity took tje place of pm** which we referred to for cme etc


Robert John Wilson, PhD

University of Tennessee HSC

Department of Radiology

865 Jefferson Ave., Chandler Bldg

Memphis, TN 38163

(p) 901-448-4274

(f) 901-448-5352

________________________________
From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu> on behalf of Jim Tomlinson via Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu>
Sent: Wednesday, April 1, 2026 10:23 AM
To: Geiser,William R <wgeiser at mdanderson.org>; Colin Paulbeck <cpaulbeck at jhu.edu>; intl_dxmedphys_wd_osu_list at lists.osu.edu <intl_dxmedphys_wd_osu_list at lists.osu.edu>
Subject: [Ext] Re: [Intl_DxMedPhys] current MQSA policy on mammo RWS MEE testing

Thank you all for your responses and advice. James A. Tomlinson, M. S. , FACR Senior Vice President Diagnostic Radiological Physicist, DABR Medical Physics Consultants, Inc. 214 E. Huron St, Ann Arbor, MI 48104 734. 662. 3197 (w) 810. 571. 0319 (c)

Thank you all for your responses and advice.



James A. Tomlinson, M.S., FACR

Senior Vice President

Diagnostic Radiological Physicist, DABR

Medical Physics Consultants, Inc.

214 E. Huron St, Ann Arbor, MI 48104

734.662.3197 (w)  810.571.0319 (c)







From: Geiser,William R [mailto:wgeiser at mdanderson.org<mailto:wgeiser at mdanderson.org>]
Sent: Wednesday, April 1, 2026 10:57 AM
To: Colin Paulbeck; Jim Tomlinson; intl_dxmedphys_wd_osu_list at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>
Subject: RE: [Intl_DxMedPhys] current MQSA policy on mammo RWS MEE testing



That list that is highlighted has been superseded by a new table that can be found in the FAQ’s for the ACR DM QC manual.



Link to the FAQ’s:  ACR Accredita<https://urldefense.com/v3/__https://edge.sitecorecloud.io/americancoldf5f-acrorgf92a-productioncb02-3650/media/ACR/Files/Clinical/Quality-Control-Manuals/Mammography-QC-Manual-FAQs.pdf__;!!KGKeukY!xWJGHcI9uqRCpLvV2TKOvZrnTLYh_L46TpqLuwWyBxae3IRZ9V9xnrkmHywv_J0_sAPxYsnx7TbnaBA19P3QE9feuzNQcfaXbB-6Bsj9zjylSA$>





[cid:image004.png at 01DCC1C9.E7716D70]



If you are not using the ACR DM QC manual then you must follow the QC manual requirements from the mammography unit manufacturer or the workstation manufacturer. See the FDA Small Entity Compliance Guide for more information. See Question 9 on Page 46 under Section O. Quality Control.



9. A facility recently had a part repaired/replaced on their mammography unit (and/or display device) and is not sure if it’s a major component that would require an MEE. What does FDA recommend in this situation?

Facilities with digital mammography units must follow a quality assurance program that is substantially the same as the quality assurance program recommended by the image receptor manufacturer (mammography unit), except that the maximum allowable dose shall not exceed the maximum allowable dose for screen-film systems in 21 CFR 900.12(e)(5)(vi). (21 CFR 900.12(e)(6)). If it is unclear whether an MEE is indicated under the image receptor manufacturer’s recommended quality assurance program for an adjustment, change, or repair of a digital mammography unit (or display device), the FDA recommends that the facility or its medical



Mammography Quality Standards Act and Regulation Amendments: Small Entity Compliance Guide - Guidance for Industry and Food and Drug Administration Staff<https://urldefense.com/v3/__https://www.fda.gov/media/181152/download__;!!KGKeukY!xWJGHcI9uqRCpLvV2TKOvZrnTLYh_L46TpqLuwWyBxae3IRZ9V9xnrkmHywv_J0_sAPxYsnx7TbnaBA19P3QE9feuzNQcfaXbB-6BsirFgI9Fw$>





Bill Geiser, MS DABR FAAPM FACR

Senior Medical Physicist

Dept. of Imaging Physics

MD Anderson Cancer Center

Houston, TX 77030

Work Cell: 832-596-3158



[New Anderson Logo]









From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu>> On Behalf Of Colin Paulbeck via Intl_dxmedphys_wd_osu_list
Sent: Wednesday, April 1, 2026 9:09 AM
To: Jim Tomlinson <jtomlinson at mpcphysics.com<mailto:jtomlinson at mpcphysics.com>>; intl_dxmedphys_wd_osu_list at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>
Subject: [EXTERNAL] Re: [Intl_DxMedPhys] current MQSA policy on mammo RWS MEE testing



Hi James, I just went through this as we changed our breast PACS vendor which required a change in the computer tower that all the monitors were plugged into. I treated this situation as a “new installation”, and did monitor testing before the

SLOW DOWN! - EXTERNAL SENDER: intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu>

Be suspicious of tone, urgency, and formatting. Do not click/open links or attachments on a mobile device. Wait until you are at a computer to confirm you are absolutely certain it is a trusted source.
If you are at all uncertain use the Report Phish button and our Cybersecurity team will investigate.

Hi James,



I just went through this as we changed our breast PACS vendor which required a change in the computer tower that all the monitors were plugged into. I treated this situation as a “new installation”, and did monitor testing before the radiologists started viewing images on them. It turned out to be pretty valuable, as some of the monitors weren’t properly set up by our IT team. It was all simple issues, like an issue with improper Windows scaling (should be 100%), or Windows not correctly identifying certain monitors. So even though you’d be reloading the same software, if it’s a fresh OS install, I’d say it’s best practice to treat it as a new installation. You’ll also want to make sure the QA Web (or whichever auto QA program is used for that monitor) was properly installed and the QA tests are all running on their schedule.



I was following the ACR QC Manual for all this. I did look somewhat thoroughly through the MQSA documentation and couldn’t find anything specifically about monitors and video cards. Here’s a screenshot of the relevant section from the ACR Manual.



[cid:image006.png at 01DCC1C9.E7716D70]



Colin



________________________________

Colin Paulbeck, PhD, DABR

Diagnostic Medical Physicist

Johns Hopkins Medicine, Baltimore MD

Office: 410-955-4181



From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu>> On Behalf Of Jim Tomlinson via Intl_dxmedphys_wd_osu_list
Sent: Tuesday, March 31, 2026 6:02 PM
To: intl_dxmedphys_wd_osu_list at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>
Subject: [Intl_DxMedPhys] current MQSA policy on mammo RWS MEE testing



      External Email - Use Caution





It’s clear to me that when a mammo RWS image display is replaced, an MEE is required before reading mammograms. But there’s been differing opinions about when the other parts of the RWS are replaced, like the tower/desktop computer. What is

It’s clear to me that when a mammo RWS image display is replaced, an MEE is required before reading mammograms.  But there’s been differing opinions about when the other parts of the RWS are replaced, like the tower/desktop computer.

What is the current MQSA policy on this?  I don’t imagine that the power supply or motherboard would affect the image quality, but are there specific features of video cards that require testing?  What if the computer is replaced and the same software is reloaded?

What policy are most of you following?



James A. Tomlinson, M.S., FACR

Senior Vice President

Diagnostic Radiological Physicist, DABR

Medical Physics Consultants, Inc.

214 E. Huron Street, Ann Arbor, MI 48104

734.662.3197 (w) 810.571.0319 (c)



The information contained in this e-mail message may be privileged, confidential, and/or protected from disclosure. This e-mail message may contain protected health information (PHI); dissemination of PHI should comply with applicable federal and state laws. If you are not the intended recipient, or an authorized representative of the intended recipient, any further review, disclosure, use, dissemination, distribution, or copying of this message or any attachment (or the information contained therein) is strictly prohibited. If you think that you have received this e-mail message in error, please notify the sender by return e-mail and delete all references to it and its contents from your systems.
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://lists.osu.edu/pipermail/intl_dxmedphys_wd_osu_list/attachments/20260401/26480bd1/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image004.png
Type: image/png
Size: 153234 bytes
Desc: image004.png
URL: <http://lists.osu.edu/pipermail/intl_dxmedphys_wd_osu_list/attachments/20260401/26480bd1/attachment.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image005.png
Type: image/png
Size: 12864 bytes
Desc: image005.png
URL: <http://lists.osu.edu/pipermail/intl_dxmedphys_wd_osu_list/attachments/20260401/26480bd1/attachment-0001.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image006.png
Type: image/png
Size: 171155 bytes
Desc: image006.png
URL: <http://lists.osu.edu/pipermail/intl_dxmedphys_wd_osu_list/attachments/20260401/26480bd1/attachment-0002.png>


More information about the Intl_dxmedphys_wd_osu_list mailing list