[Intl_DxMedPhys] CT Fluoroscopy Testing Recommendations
Szczykutowicz, Timothy P
TSzczykutowicz at uwhealth.org
Wed Mar 11 14:13:31 EDT 2026
Testing IQ, scatter dose, CTDIvol?
That mode has some different set ups possible. We use it every day. Physicians should be only quickly tapping on the pedal so the tube does a short scan. If they are pressing the pedal, the tube will stay on. You can compare the total acquisition time to the rotation time. If they are using the mode correctly, the beam on time should ALWAYS be less than the rotation time. If they say they need real time ct fluoro they don't.
I attached a paper we wrote which lets you calculate the physician scatter dose based on technique factors.
Below is a plot of image vs time, for CT fluoro, if the rad has a lead foot and is doing real time CTF you get "flat spots" on a plot like this,
I can monitor this now by pulling the CTF cases. And to use for rad/resident/fellow training. I show so you can see how you could also monitor this, even if you dont have ability to pull images from pacs, you can just go to pacs and look at image acquisition time tags in a ctf run to check for this behavior by rads.
[cid:d5e98141-1a3a-44c3-91e4-f08ae820ddfb]
Timothy P. Szczykutowicz, Ph.D., DABR
Professor
Departments of Radiology, Medical Physics and BME
University of Wisconsin Madison
Cell# 1-716-560-7751<tel:(716)%20560-7751>
Office# 1-608-263-5729
he/him/his
________________________________
From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces+tszczykutowicz=uwhealth.org at lists.osu.edu> on behalf of Holmes, John via Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu>
Sent: Tuesday, March 10, 2026 10:19 AM
To: intl_dxmedphys_wd_osu_list at lists.osu.edu <intl_dxmedphys_wd_osu_list at lists.osu.edu>
Subject: [Intl_DxMedPhys] CT Fluoroscopy Testing Recommendations
WARNING: This email appears to have originated outside of the UW Health email system.
DO NOT CLICK on links or attachments unless you recognize the sender and know the content is safe.
Good Morning Everyone, Does anyone have recommendations or personal experiences with testing the fluoroscopy (smartstep on GE VCTs) mode of CT scanners? We are investigating personnel exposures in conjunction with the scanner’s procedural
Good Morning Everyone,
Does anyone have recommendations or personal experiences with testing the fluoroscopy (smartstep on GE VCTs) mode of CT scanners?
We are investigating personnel exposures in conjunction with the scanner’s procedural use and are testing beyond the usual state requirements, ACR requirements, and CT fluoro CTDI measurements.
John
John R. Holmes, DMP, DABR
Diagnostic Medical Physicist
Program Director: Diagnostic Medical Physics Residency
Assistant Professor: Radiology
University of Pittsburgh
Office of Research Protections: Radiation Safety
Hieber Building, Suite 400
3500 5th Ave.
Pittsburgh, PA 15213
412-624-2725
jrh167 at pitt.edu<mailto:jrh167 at pitt.edu>
pitt.edu<https://urldefense.com/v3/__http://www.pitt.edu/__;!!KGKeukY!0uREaWgBgeMma-eYU6BXXiiBrC4b3MWGVF4dh6Sv9UtlUdb8skTGrVruzGMg5y__o7jVA4n2d9m0oWt0T7TSiF0RC9yHCSem8mdvTg$>
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://lists.osu.edu/pipermail/intl_dxmedphys_wd_osu_list/attachments/20260311/9ac4d7f2/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image.png
Type: image/png
Size: 121868 bytes
Desc: image.png
URL: <http://lists.osu.edu/pipermail/intl_dxmedphys_wd_osu_list/attachments/20260311/9ac4d7f2/attachment.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: J Applied Clin Med Phys - 2024 - Knott - Physician scatter dose in interventional CT fluoroscopy.pdf
Type: application/pdf
Size: 733471 bytes
Desc: J Applied Clin Med Phys - 2024 - Knott - Physician scatter dose in interventional CT fluoroscopy.pdf
URL: <http://lists.osu.edu/pipermail/intl_dxmedphys_wd_osu_list/attachments/20260311/9ac4d7f2/attachment.pdf>
More information about the Intl_dxmedphys_wd_osu_list
mailing list