[Intl_DxMedPhys] GE Pristina Flat Field Uniformity
Travis White
Travis.White at PrismaHealth.org
Tue Jan 27 09:26:58 EST 2026
I have seen a similar issue on GE units, usually after a service visit. The working theory at the time was that that particular calibration has been done late in the calibration process, when the engineer was rushing and the system was hot. The filters had thermally expanded when the calibration was done, then cooled and resulted in the “orange peel” artifact when tested under cooler (normal) conditions. I would have the FSE come back and redo that particular calibration and it would be fine after that. Not sure if this is your issue, but if you had recent service, it might be worth asking at what point in the calibration process that particular target/filter combo is performed.
From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu> On Behalf Of Palmer, Matthew R. (BIDMC - Radiology) via Intl_dxmedphys_wd_osu_list
Sent: Monday, January 26, 2026 4:39 PM
To: DxMedPhys List <intl_dxmedphys_wd_osu_list at lists.osu.edu>
Subject: [Intl_DxMedPhys] GE Pristina Flat Field Uniformity
Dear Colleagues, Multiple times recently, I’ve found artifacts when performing the Artifact Eval and Flat Field uniformity test, on the 2D Rh/Ag acquisitions for the GE Pristina mammo unit. I’ve pasted in a portion of my iPhone snapshot
ZjQcmQRYFpfptBannerStart
CAUTION! **EXTERNAL EMAIL**
Do NOT click links or open attachments unless you recognize the sender and know the content is safe.
Report Suspicious <https://us-phishalarm-ewt.proofpoint.com/EWT/v1/ClysJhrlK38!de2C7i07uHPwhQfTYWbOKhZxgO3pWCinrhXbBNxQfl7kc7ithKK_ScS-mqCHwVYvsr60rvXzYMOG7N2VsDtrXDz5iTicgt7TZY83lEXnyLku4uj63tLiYbg-PqeBdULqrBgZ14ZZ0gro7EYJKwmw$ >
ZjQcmQRYFpfptBannerEnd
Dear Colleagues, Multiple times recently, I’ve found artifacts when performing the Artifact Eval and Flat Field uniformity test, on the 2D Rh/Ag acquisitions for the GE Pristina mammo unit. I’ve pasted in a portion of my iPhone snapshot
Dear Colleagues,
Multiple times recently, I’ve found artifacts when performing the Artifact Eval and Flat Field uniformity test, on the 2D Rh/Ag acquisitions for the GE Pristina mammo unit. I’ve pasted in a portion of my iPhone snapshot from the AWS display. The image quality isn’t pristine(a) and I’ve introduced a Moire pattern (there’s no ‘zipper’-like artifact in reality) but I think you can see the mottled appearance pretty clearly.
After the FSE runs the gain cal, the artifacts disappear – for about a day then reappear.
FSE’s feedback from GE experts is that not to worry – the artifacts won’t affect clinical images. That’s hard to swallow.
I was astonished that the tech’s weren’t picking up the artifacts in their weekly QC so I ran their Flat Field test from the tech IQST Weekly QC page. The images shown briefly (but not saved as far as I can tell) during the test sequence look perfect (although you can’t adjust the window/level).
I’m wondering if anyone has had this problem or can shed any light on it.
Matt Palmer
BIDMC Radiology
Matthew R. Palmer, PhD, DABR
Director, Medical Imaging Physics, Department of Radiology
Assistant Professor, Harvard Medical School
Beth Israel Deaconess Medical Center
330 Brookline Avenue
Boston, MA 02215
office: 617-667-0176 fax: 617-667-2770
mobile: 617-794-1736 pager: 617-632-7243 #32214
________________________________
PLEASE NOTE: This message is intended for the use of the person to whom it is addressed. It may contain information that is privileged, confidential and exempt from disclosure under applicable law. If you are not the intended recipient, your use of this message for any purpose is strictly prohibited. If you have received this communication in error, please delete the message and notify the sender so that we may correct our records. See our web page at https://urldefense.com/v3/__http://www.bilh.org__;!!KGKeukY!01JoZKxbyLbupN8rasp6fnPqt4fq-I0ZLsNfZZZTKg65VoE8Or5Tzevp73S8JNYlIdXBgXO5L6u9tYHOKEvsToKfFWvnZLkfIZYvwZzFw6EgKzGS9g$ <https://urldefense.com/v3/__http:/www.bilh.org__;!!KGKeukY!yPXjQjwtVfXFt-7mTxRycQxsuL5wT3jVEnZVQtVDhgKLkcWAtRAcVpzHasCOwejFq3t4iE207kdiJ4S7kfrqV3zLI7pq7IHjXHYpMUWnz72j2EY$> for a full directory of Beth Israel Lahey Health sites, staff, services and career opportunities.
----------------------------------------------------------------------
CONFIDENTIALITY NOTICE
This message is intended only for the use of the individual or entity to which it is addressed and may contain information that is privileged, confidential and exempt from disclosure under applicable law. If the reader of this message is not the intended recipient, you are hereby notified that any dissemination, distribution, or copying of this communication is strictly prohibited by law. If you have received this communication in error, please notify me immediately.
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20260127/e4e77631/attachment.html>
More information about the Intl_dxmedphys_wd_osu_list
mailing list