Hi Stephanie,
I made some enquiries to an applications specialist in Australia last year and received the below response.
------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------
It is not quite a rebrand, as the design is to make it applicable and transferable between different scanners (including other models). i.e. you set the CARE IQ level on the X.cite at Cairns, and if the protocol is imported onto the go.Top at Atherton then the underlying Qref. mAs etc. Will adapt to provide consistent image quality for standarisation.
There is not a lot of information in the IFU/System owner manual, though luckily my Applications Colleagues handled a similar enquiry recently from a Victorian Physicist. The answer from R&D is below:
1.)
As described in the IFU, CARE kV IQ level is proportional to Q. ref. mAs at 120 kV. The proportionality factor depends on the scanner model type and takes IQ relevant factors (such as the gantry bore size) into account. It is designed to provide consistent image quality over different scanner models if the same scan protocol with the same IQ level is used. In exam designer, the proportionality factor can be seen by comparing IQ level and Q. ref. mAs for a scan range with 120 kV.
2.)
In most cases, the CTDIvol is adapted to the Som/7 scanner so that the customer has the same or better IQ. However, this must be discussed with the customer and is the job of the application specialist.
Within Som/10 scanners, it is no longer a problem at all to transfer the protocols. So in this case, adapt the protocols once from Som/7 to Som/10 and export/import to the other Som/10 scanners.
Keep in mind, we recommend starting with the 'largest' scanner.
There is no general table to provide for the parameters when coming from SOM7 to SOM10. It is depending on with which image quality the customer is satisfied with.
------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------
I didn’t find their response overly helpful so I remain optimistic that further documentation will be released in the future.
Kind Regards,
[Queensland Government]
Gene Jackson
Medical Physicist
Service Integration and Biomedical Technology Branch
Townsville University Hospital
Clinical Support Queensland | Queensland Health
P
(07) 4433 2244<tel:(07)%204433%201152> | 0417 902 041<tel:0417%20902%20041>
E
Gene.Jackson at health.qld.gov.au<mailto:Gene.Jackson at health.qld.gov.au>
W
health.qld.gov.au<https://urldefense.com/v3/__https://www.health.qld.gov.au/__;!!KGKeukY!2XptRZV4TIRu73HSL9zyX6MAF4xOwsCr3GKhUi9QQbxQQ5Q8OqcvLaGbF_CPCtxEnL3sdGFwVw6BeRKwmujg9RNa6nwc6Zan2ye373H9g9qQ3n-9_Y2xRZQ$ >
A
Level 1, Acute Block, 100 Angus Smith Drive, Douglas, QLD 4814<https://urldefense.com/v3/__https://www.health.qld.gov.au/__;!!KGKeukY!2XptRZV4TIRu73HSL9zyX6MAF4xOwsCr3GKhUi9QQbxQQ5Q8OqcvLaGbF_CPCtxEnL3sdGFwVw6BeRKwmujg9RNa6nwc6Zan2ye373H9g9qQ3n-9_Y2xRZQ$ >
[Delivering for Queensland]
[facebook]<https://urldefense.com/v3/__https://www.facebook.com/QLDHealth/?fref=nf__;!!KGKeukY!2XptRZV4TIRu73HSL9zyX6MAF4xOwsCr3GKhUi9QQbxQQ5Q8OqcvLaGbF_CPCtxEnL3sdGFwVw6BeRKwmujg9RNa6nwc6Zan2ye373H9g9qQ3n-92vNBV6g$ >
[twitter]<https://urldefense.com/v3/__https://twitter.com/qldhealthnews__;!!KGKeukY!2XptRZV4TIRu73HSL9zyX6MAF4xOwsCr3GKhUi9QQbxQQ5Q8OqcvLaGbF_CPCtxEnL3sdGFwVw6BeRKwmujg9RNa6nwc6Zan2ye373H9g9qQ3n-94He7WoE$ >
[linkedin]<https://urldefense.com/v3/__http://www.linkedin.com/company/queensland-health__;!!KGKeukY!2XptRZV4TIRu73HSL9zyX6MAF4xOwsCr3GKhUi9QQbxQQ5Q8OqcvLaGbF_CPCtxEnL3sdGFwVw6BeRKwmujg9RNa6nwc6Zan2ye373H9g9qQ3n-9e7nejwE$ >
[instagram]<https://urldefense.com/v3/__https://www.instagram.com/queenslandhealth/__;!!KGKeukY!2XptRZV4TIRu73HSL9zyX6MAF4xOwsCr3GKhUi9QQbxQQ5Q8OqcvLaGbF_CPCtxEnL3sdGFwVw6BeRKwmujg9RNa6nwc6Zan2ye373H9g9qQ3n-9Zb_-XEs$ >
From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces+gene.jackson=health.qld.gov.au at lists.osu.edu> On Behalf Of Stephanie M. Leon via Intl_dxmedphys_wd_osu_list
Sent: Friday, 9 January 2026 5:20 AM
To: Stephanie M. Leon <Stephanie.M.Leon at hitchcock.org>; intl_dxmedphys_wd_osu_list at lists.osu.edu
Subject: Re: [Intl_DxMedPhys] Siemens CT IQ level
Just to clarify, if it makes a difference, this is not a PCCT scanner. It’s a Pro. Pulse and I’m wondering about a traditional single-energy scan. From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces@ lists. osu. edu> On Behalf
This email originated from outside Queensland Health. DO NOT click on any links or open attachments unless you recognise the sender and know the content is safe.
________________________________
Just to clarify, if it makes a difference, this is not a PCCT scanner. It’s a Pro.Pulse and I’m wondering about a traditional single-energy scan.
From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu>> On Behalf Of Stephanie M. Leon via Intl_dxmedphys_wd_osu_list
Sent: Thursday, January 8, 2026 2:04 PM
To: intl_dxmedphys_wd_osu_list at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>
Subject: [Intl_DxMedPhys] Siemens CT IQ level **EXTERNAL**
Hi all, Apparently the newer Siemens CT scanners have dispensed with reference mAs and now use IQ (image quality) Level. I’m having trouble finding any detail about IQ Level. Is there a way to directly relate it to the old reference mAs,
Hi all, Apparently the newer Siemens CT scanners have dispensed with reference mAs and now use IQ (image quality) Level. I’m having trouble finding any detail about IQ Level. Is there a way to directly relate it to the old reference mAs,
Hi all,
Apparently the newer Siemens CT scanners have dispensed with reference mAs and now use IQ (image quality) Level. I’m having trouble finding any detail about IQ Level. Is there a way to directly relate it to the old reference mAs, assuming a single-energy scan at the same kVp?
Thanks!
Stephanie Leon
IMPORTANT NOTICE REGARDING THIS ELECTRONIC MESSAGE:
This message is intended for the use of the person to whom it is addressed and may contain information that is privileged, confidential, and protected from disclosure under applicable law. If you are not the intended recipient, your use of this message for any purpose is strictly prohibited. If you have received this communication in error, please delete the message and notify the sender so that we may correct our records.
IMPORTANT NOTICE REGARDING THIS ELECTRONIC MESSAGE:
This message is intended for the use of the person to whom it is addressed and may contain information that is privileged, confidential, and protected from disclosure under applicable law. If you are not the intended recipient, your use of this message for any purpose is strictly prohibited. If you have received this communication in error, please delete the message and notify the sender so that we may correct our records.
**********************************************************************************
Disclaimer: This email and any attachments may contain legally privileged or confidential information and may be protected by copyright. You must not use or disclose them other than for the purposes for which they were supplied. The privilege or confidentiality attached to this message and attachments is not waived by reason of mistaken delivery to you. If you are not the intended recipient, you must not use, disclose, retain, forward or reproduce this message or any attachments. If you receive this message in error, please notify the sender by return email or telephone and destroy and delete all copies. Unless stated otherwise, this email represents only the views of the sender and not the views of the Queensland Government.
Queensland Health carries out monitoring, scanning and blocking of emails and attachments sent from or to addresses within Queensland Health for the purposes of operating, protecting, maintaining and ensuring appropriate use of its computer network.
**********************************************************************************
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20260109/113b5107/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image001.png
Type: image/png
Size: 16238 bytes
Desc: image001.png
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20260109/113b5107/attachment.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image002.png
Type: image/png
Size: 33853 bytes
Desc: image002.png
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20260109/113b5107/attachment-0001.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image003.png
Type: image/png
Size: 826 bytes
Desc: image003.png
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20260109/113b5107/attachment-0002.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image004.png
Type: image/png
Size: 1075 bytes
Desc: image004.png
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20260109/113b5107/attachment-0003.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image005.png
Type: image/png
Size: 923 bytes
Desc: image005.png
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20260109/113b5107/attachment-0004.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image006.png
Type: image/png
Size: 1303 bytes
Desc: image006.png
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20260109/113b5107/attachment-0005.png>
More information about the Intl_dxmedphys_wd_osu_list
mailing list