Since Angio cases will take longer and scatter less, is it reasonable to assume that *any* amount of angio will reduce the required lead rather than increase it?
If so, then the worst case scenario is going to be assuming 100% CT workload and 0% angio, so the shielding design could be done simply. Also, if they ditch this room in 10 yrs for a CT or for a regular angio, the shielding should be fine as is.
---
Thanks,
Gary
On Mon, 29 Dec 2025 13:38:07 -0600, Pavlina Boxx via Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu> wrote:
>Hi All,
>
>Our hospital has purchased a Siemens Nexaris Angio-CT and I am getting
>ready to start working on the shielding plan. If you have one of
>these, can you please share an approximate number of CT vs Angio exams
>done in a week. Thank you,
>
> Pavlina
>