 Hi Bob,
I don't recall a direct ACR reference for this policy but do recall discussions among the ACR MRI Technical Subcommittee members that no changes to a protocol should be allowed that were not also applied to normal clinical imaging.  To do so would gut the basis of MRI performance testing by allowing technical results that could not be accomplished clinically.
Features of the initial ACR MRI phantom were designed to ensure fair testing by eliminating ways one could bias results.  For example, slice thickness, resolution and bandwidth (indirectly from the fat/water inserts) are all assessed simultaneously in the ACR T1 axial series.  One can't artificially increase SNR by using a wider slice thickness, by reducing resolution via a lower matrix or abnormal k-space weighting, or by markedly lowering bandwidth.  I also seem to recall ACR noting not to chill the large ACR MRI phantom prior to testing in order to improve SNR as some did in the early days of low-field systems. 
Bob Bell760-579-3929  


    On Monday, November 10, 2025 at 08:07:02 PM PST, Bob Kobistek <bob at rjkmedphys.com> wrote:  
 
 
Is that documented somewhere on the ACR website?
 
  
 
  
 
Robert J. Kobistek, MS, FACR, DABR, MRSE(MRSC™)
 
Medical Physicist
 
RJKMedical Physics, Inc.
 
440-463-7879
 
  
 
From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces+bob=rjkmedphys.com at lists.osu.edu>On Behalf Of Bob Bell via Intl_dxmedphys_wd_osu_list
Sent: Monday, November 10, 2025 8:01 PM
To: Hardy, Peter A. <peter.hardy at uky.edu>; Gretchen Raterman Bell <gretchen.raterman at gmail.com>
Cc: kenharper14 via Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu>
Subject: Re: [Intl_DxMedPhys] Siemens MRI ACR protocol
 
  
 
Hi all, Just a quick comment on this email chain. It's my understanding ACR allows any filters or instrument settings routinely used for patient studies to be used for phantom studies but does not allow filters/settings unique to phantom studies
 
Hi all,
 
  
 
Just a quick comment on this email chain.  It's my understanding ACR allows any filters or instrument settings routinely used for patient studies to be used for phantom studies but does not allow filters/settings unique to phantom studies alone.  Employing filters/settings that are not used for patient studies defeats the purpose of annual testing by artificially altering instrument performance from normal clinical practice.  For example, it's okay to use prescan normalize on Siemens units to improve IIU as long as it is also used for patient studies.  However, altering filters/settings to those not common for clinical studies just to improve instrument phantom performance is a no no.  
 
  
 
Bob Bell
 
  
 
On Friday, November 7, 2025 at 06:29:18 AM PST, Gretchen Raterman Bell via Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu> wrote: 
 
  
 
  
 
For what it's worth, I'll summarize some settings I've gotten from both the DXlist and from Siemens that have helped tremendously. I work on both Aeras and Solas. To improve signal uniformity: - Use B1 filter - Use sum of squares
 
For what it's worth, I'll summarize some settings I've gotten from both the DXlist and from Siemens that have helped tremendously.  I work on both Aeras and Solas.
 
  
 
To improve signal uniformity:
 
- Use B1 filter
 
- Use sum of squares (coil combination)
 
  
 
To improve spatial resolution, particularly Solas:
    
   - Resolution tab àB1 Filter should be ON
   - Elliptical Filter should be OFF
   - Prescan Normalize should be ON
   - Sequence tab àGradient Mode should be NORMAL or FAST
   - Sequence tab à RF pulse types, try different SAR settings (try this very last, he was unsure about this one)
   - If none of this improves the high contrast resolution, he highly recommended calling into the  1800 number for remote service:
    
   
   - In-Line Image Filter should be turned OFF
   - This is a global setting that only FSE or remote service can modify; must be turned back on after acquiring phantom images
   - It tends to be an overall smoothing filter
   - Particularly affects newer software, which is likely why we don’t see it on the Aeras
 
  
 
FYI, I haven't had to do that last bullet point yet, but good to know just in case.
 
Gretchen R. Bell, M.S., DABR
Diagnostic Imaging Physicist
Ochsner Medical Center
(504)842-8506
 
  
 
  
 
On Fri, Nov 7, 2025 at 8:12 AM Hardy, Peter A. via Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu> wrote:
 

It might be easier to import an .exar file. These are available from Siemens on their Phoenix site. ACR Phantom QA Test I’m attaching the file for a 1. 5T system. Similar files are available for other systems and OS levels. This protocol likely
 
It might be easier to import an .exar file. These are available from Siemens on their Phoenix site.
 
ACR Phantom QA Test
 
 
 
I’m attaching the file for a 1.5T system. Similar files are available for other systems and OS levels.
 
This protocol likely includes the updated axial T2 sequence.
 
 
 
Peter Hardy, PhD, DABMP
 
Associate Professor of Radiology, Neuroscience and Biomedical Engineering
 
Faculty, Division of Diagnostic & Nuclear Medical Physics
 
Department of Radiology
 
University of Kentucky College of Medicine
 
UK HealthCare
 
800 Rose Street, Room HX-307
 
Lexington, KY 40536-0293
 
Phone: (859) 323-2954
 
Email:peter.hardy at uky.edu
 
 
 
 
 
From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces+peter.hardy=uky.edu at lists.osu.edu>On Behalf Of Adam Springer via Intl_dxmedphys_wd_osu_list
Sent: Friday, November 7, 2025 7:26 AM
To: kenharper14 via Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu>
Subject: [Intl_DxMedPhys] Siemens MRI ACR protocol
 
 
 
CAUTION: External Sender
 
 
 
Hello, List. I have instructions from Siemens for building the ACR protocol on a Magntom Avanto, from 2005. Does anyone have a more recent set of those instructions? I've had to ask Siemens to reload their ACR protocol at a couple sites recently,
 
Hello, List.
 
 
 
I have instructions from Siemens for building the ACR protocol on a Magntom Avanto, from 2005. Does anyone have a more recent set of those instructions? I've had to ask Siemens to reload their ACR protocol at a couple sites recently, and it would be convenient if I could just change the settings while I'm on site.
 
 
 
Happy Friday!
 
Adam
 
 
 
 
 
 
 
========================
Adam C. Springer, MS, DABR®
Medical Physicist & CEO
KLS Physics Group, LLC
124 Killgore Rd
Ruston, LA 71270
Office: 318-255-5033
Mobile: 318-272-8292
 

 
  
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20251111/3ba95cf7/attachment.html>


More information about the Intl_dxmedphys_wd_osu_list mailing list