For what it's worth, I'll summarize some settings I've gotten from both the
DXlist and from Siemens that have helped tremendously.  I work on both
Aeras and Solas.

To improve signal uniformity:
- Use B1 filter
- Use sum of squares (coil combination)

To improve spatial resolution, particularly Solas:

   - Resolution tab à B1 Filter should be ON
   - Elliptical Filter should be OFF
   - Prescan Normalize should be ON
   - Sequence tab à Gradient Mode should be NORMAL or FAST
   - Sequence tab à RF pulse types, try different SAR settings (try this
   very last, he was unsure about this one)
   - If none of this improves the high contrast resolution, he highly
   recommended calling into the  1800 number for remote service:
      - In-Line Image Filter should be turned OFF
      - This is a global setting that only FSE or remote service can
      modify; must be turned back on after acquiring phantom images
      - It tends to be an overall smoothing filter
      - Particularly affects newer software, which is likely why we don’t
      see it on the Aeras


FYI, I haven't had to do that last bullet point yet, but good to know just
in case.

Gretchen R. Bell, M.S., DABR
Diagnostic Imaging Physicist
Ochsner Medical Center
(504)842-8506



On Fri, Nov 7, 2025 at 8:12 AM Hardy, Peter A. via
Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu> wrote:

> It might be easier to import an .exar file. These are available from
> Siemens on their Phoenix site. ACR Phantom QA Test I’m attaching the file
> for a 1. 5T system. Similar files are available for other systems and OS
> levels. This protocol likely
>
> It might be easier to import an .exar file. These are available from
> Siemens on their Phoenix site.
>
> ACR Phantom QA Test
> <https://urldefense.com/v3/__https://www.magnetomworld.siemens-healthineers.com/clinical-corner/protocols/mr-in-rt/acr-phantom-qa-test__;!!KGKeukY!0J5pWek2Y1lqkWdNUarukJAaQxlT5NXRKlh0kmDROFkR-peOYmfS1MRNkentpqiJPmhBL1nfqqpvJzIqMmXyItg3OYVYBOFj01ry3nznJwbc$>
>
>
>
> I’m attaching the file for a 1.5T system. Similar files are available for
> other systems and OS levels.
>
> This protocol likely includes the updated axial T2 sequence.
>
>
>
> *Peter Hardy, PhD, DABMP*
>
> Associate Professor of Radiology, Neuroscience and Biomedical Engineering
>
> Faculty, Division of Diagnostic & Nuclear Medical Physics
>
> Department of Radiology
>
> University of Kentucky College of Medicine
>
> UK HealthCare
>
> 800 Rose Street, Room HX-307
>
> Lexington, KY 40536-0293
>
> Phone: (859) 323-2954
>
> Email: peter.hardy at uky.edu
>
>
>
>
>
> *From:* Intl_dxmedphys_wd_osu_list
> <intl_dxmedphys_wd_osu_list-bounces+peter.hardy=uky.edu at lists.osu.edu> *On
> Behalf Of *Adam Springer via Intl_dxmedphys_wd_osu_list
> *Sent:* Friday, November 7, 2025 7:26 AM
> *To:* kenharper14 via Intl_dxmedphys_wd_osu_list <
> intl_dxmedphys_wd_osu_list at lists.osu.edu>
> *Subject:* [Intl_DxMedPhys] Siemens MRI ACR protocol
>
>
>
> *CAUTION:** External Sender*
>
>
>
> Hello, List. I have instructions from Siemens for building the ACR
> protocol on a Magntom Avanto, from 2005. Does anyone have a more recent set
> of those instructions? I've had to ask Siemens to reload their ACR protocol
> at a couple sites recently,
>
> Hello, List.
>
>
>
> I have instructions from Siemens for building the ACR protocol on a
> Magntom Avanto, from 2005. Does anyone have a more recent set of those
> instructions? I've had to ask Siemens to reload their ACR protocol at a
> couple sites recently, and it would be convenient if I could just change
> the settings while I'm on site.
>
>
>
> Happy Friday!
>
> Adam
>
>
>
>
>
>
>
> ========================
> Adam C. Springer, MS, DABR®
> Medical Physicist & CEO
> KLS Physics Group, LLC
> 124 Killgore Rd
> Ruston, LA 71270
> Office: 318-255-5033
> Mobile: 318-272-8292
>
> [image: 2927183d-095e-413d-96b1-930240ed6257]
>
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20251107/eeaa9fd3/attachment.html>


More information about the Intl_dxmedphys_wd_osu_list mailing list