It’s 30cm for both mobile and stationary units in Europe. So I suppose it makes sense that it’s Siemens. They already make the 30cm covers, probably a case of “why not offer it”.
From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces+teresa.lo=act.gov.au at lists.osu.edu> On Behalf Of Tao, Ashley T via Intl_dxmedphys_wd_osu_list
Sent: Friday, 17 October 2025 5:24 AM
To: DxMedPhys List <intl_dxmedphys_wd_osu_list at lists.osu.edu> <intl_dxmedphys_wd_osu_list at lists.osu.edu>; William Sensakovic <wfsensak at outlook.com>
Subject: Re: [Intl_DxMedPhys] Fluoro minimum SSD
Caution: This email originated from outside of the ACT Government. Do not click links or open attachments unless you recognise the sender and know the content is safe.
Its the Siemens Artis Icono. Still awaiting a response from them. From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces+attao=gundersenhealth. org@ lists. osu. edu> on behalf of William Sensakovic via Intl_dxmedphys_wd_osu_list
Its the Siemens Artis Icono. Still awaiting a response from them.
________________________________
From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces+attao=gundersenhealth.org at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list-bounces+attao=gundersenhealth.org at lists.osu.edu>> on behalf of William Sensakovic via Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>>
Sent: Thursday, October 16, 2025 1:19 PM
To: DxMedPhys List <intl_dxmedphys_wd_osu_list at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>> <intl_dxmedphys_wd_osu_list at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>>
Subject: Re: [Intl_DxMedPhys] Fluoro minimum SSD
The cfr seems pretty clear that 30cm is only for portables. Are you comfortable sharing the vendor? Small and new vendors often save money by not hiring experts in informatics nor regulations. Had this experience before, though not for the cone
CAUTION: This email originated from outside of Emplify Health. Do not click on any links or open any attachments unless you recognize the sender and know the contents are safe.
We will NEVER ask you to provide your user name or password in reply to an email.
Questions? Contact the Bellin Service Desk at x258600 / (920) 436-8600 or the Gundersen Help Desk at x53070 / (608) 775-3070.
The cfr seems pretty clear that 30cm is only for portables. Are you comfortable sharing the vendor? Small and new vendors often save money by not hiring experts in informatics nor regulations. Had this experience before, though not for the cone size. They basically just want to make a mark quickly and get bought out by one of the big boys. Curious if this is an established vendor or small start up.
Best,
Bill
William F. Sensakovic, PhD, MRSC(MRSE), FAAPM
Chair and Assoc. Prof., Mayo Clinic (Arizona)
Founder, Telerad Physics Teaching, LLC
Twitter: @wfsensak
________________________________
From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu>> on behalf of Tao, Ashley T via Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>>
Sent: Thursday, October 16, 2025 11:04 AM
To: DxMedPhys List <intl_dxmedphys_wd_osu_list at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>> <intl_dxmedphys_wd_osu_list at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>>
Subject: [Intl_DxMedPhys] Fluoro minimum SSD
Hey all,
We recently had a stationary c-arm unit installed and it came with a short and tall cover. The short cover provides a minimum SSD of 30cm whereas the tall cover provides a minimum SSD of 38cm. The manufacturer said their customers typically install the short cover because that's what the providers prefer. It gives them more flexibility with angles. Sure enough, that's what our providers wanted.
Both 21 CFR and state of wisconsin says that the minimum SSD for stationary fluoroscopes must be 38 cm, but the short cover does not meet this requirement. Curious if anyone else has dealt with this. Their FSE says that the customer has the choice of which cover they want to use. I have asked for feedback from the manufacturer but have not yet received a response. Before I upset my providers by making them switch to the tall cover, is there something in the regs that I'm not seeing that allows them to use the 30 cm SSD? Would anyone know why they offer the short cover if fed and state requirements require 38 cm?
Thanks
Ashley
Ashley Tao, PhD, DABR, MRSE (MRSC™)
Lead Diagnostic Medical Physicist
Radiation Safety Officer
Emplify Health by Gundersen
1900 South Ave
Mail Stop H01-016
La Crosse, WI 54601
attao at emplifyhealth.org<mailto:attao at emplifyhealth.org>
(w) 608-775-2078
(c) 507-398-1561
[cid:image001.png at 01DC3F4C.FA55E5B0]
-----------------------------------------------------------------------
This email, and any attachments, may be confidential and also privileged. If you are not the intended recipient, please notify the sender and delete all copies of this transmission along with any attachments immediately. You should not copy or use it for any purpose, nor disclose its contents to any other person.
-----------------------------------------------------------------------
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20251016/c3d93a7d/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image001.png
Type: image/png
Size: 7641 bytes
Desc: image001.png
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20251016/c3d93a7d/attachment.png>
More information about the Intl_dxmedphys_wd_osu_list
mailing list