Hi All,

We have been using myAQS for a few years now for CT and it has been going well (Daily QC using vendor phantoms and ACR testing). We have not explored their MRI option as the Canadian requirements are different – but they do support ACR phantom analysis for CT and MRI.

The webapp checks all 4 of Joseph’s boxes below.

Kind regards,

Thor Bjarnason, PhD, PEng, FCCPM
Medical Physicist
Interior Health, Kelowna, BC
Office: (250) 980-6157 | Mobile: (250) 826-0722
Thor.Bjarnason at interiorhealth.ca



From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu> On Behalf Of Erlend Kristoffer Frivold Andersen via Intl_dxmedphys_wd_osu_list
Sent: 17 August 2025 11:53 PM
To: DxMedPhys List <intl_dxmedphys_wd_osu_list at lists.osu.edu>
Subject: Re: [Intl_DxMedPhys] Seeking Daily CT and MR QC Automated Software

Caution: This email originated from outside of Interior Health. Do not click links or open attachments unless you recognize the sender, their email address, and know the content is safe. If you suspect this is a phishing or fraudulent email please forward it to spam at interiorhealth.ca<mailto:spam at interiorhealth.ca>.

Hi, We use the WAD-QC package from NVKF (MedPhysNL / wadqc — Bitbucket). It takes a bit of work to setup, but you can write custom analysis modules that is triggered when images are sent to the dicom server. We currently use it for CT and DX,
Hi,

We use the WAD-QC package from NVKF (MedPhysNL / wadqc — Bitbucket<https://urldefense.com/v3/__https://bitbucket.org/MedPhysNL/wadqc/wiki/Home__;!!KGKeukY!xW7Lq0njVJcZl8z4I30T3PAQSSNKcVxB9BFLXugbDZp7IwENkUR82GFdebIHISTo9JA95yhm4wywj5oxHkJC2XQ6sGwEsjJD6i78fuLPMPgFGkucjw$>). It takes a bit of work to setup, but you can write custom analysis modules that is triggered when images are sent to the dicom server. We currently use it for CT and DX, the tech run daily QC and sends the images to a dicom server, triggers an analysis and results are available after a couple of minutes on a (selfhosted) website.

-Erlend

Erlend Andersen

Medisinsk fysiker | +47 482 34 642 | erlend.andersen at sshf.no<mailto:erlend.andersen at sshf.no>

[https://urldefense.com/v3/__https://sshf.no/PublishingImages/Logo/Logo_til_mail_signatur_bruk-M**A_IKKE_SLETTES.jpg__;w4U!!KGKeukY!wRsjDeKYedtrpgTMZyr_TGBgp9-HmeGdnyyJ3n1M37DK87Gj5TJiyBT6tPJ-wGpH-KuuGfwrNjbRMZ_U3nuv6vpZi7A9SKm2jqFTnisNJW8OWbHF5BZUVmM$ ]<https://urldefense.com/v3/__https://www.sshf.no/__;!!KGKeukY!xW7Lq0njVJcZl8z4I30T3PAQSSNKcVxB9BFLXugbDZp7IwENkUR82GFdebIHISTo9JA95yhm4wywj5oxHkJC2XQ6sGwEsjJD6i78fuLPMPhlwk8zlQ$>
Medisinsk serviceklinikk | Postboks 416, 4604 Kristiansand | https://urldefense.com/v3/__http://www.sshf.no__;!!KGKeukY!wRsjDeKYedtrpgTMZyr_TGBgp9-HmeGdnyyJ3n1M37DK87Gj5TJiyBT6tPJ-wGpH-KuuGfwrNjbRMZ_U3nuv6vpZi7A9SKm2jqFTnisNJW8OWbHF14ljFKU$ <https://urldefense.com/v3/__http://www.sshf.no/__;!!KGKeukY!xW7Lq0njVJcZl8z4I30T3PAQSSNKcVxB9BFLXugbDZp7IwENkUR82GFdebIHISTo9JA95yhm4wywj5oxHkJC2XQ6sGwEsjJD6i78fuLPMPhlDMnHiQ$>


From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces+erlend.andersen=sshf.no at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list-bounces+erlend.andersen=sshf.no at lists.osu.edu>> On Behalf Of Jeff Moirano via Intl_dxmedphys_wd_osu_list
Sent: Monday, August 18, 2025 8:38 AM
To: DxMedPhys List <intl_dxmedphys_wd_osu_list at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>>
Subject: Re: [Intl_DxMedPhys] Seeking Daily CT and MR QC Automated Software

I'd take a look at Atirix too, I've been impressed. They have Nuc Med. Jeff On Fri, Aug 15, 2025 at 11: 37 AM Januseski, Daniel via Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list@ lists. osu. edu> wrote: This message was sent
I'd take a look at Atirix too, I've been impressed. They have Nuc Med.

Jeff

On Fri, Aug 15, 2025 at 11:37 AM Januseski, Daniel via Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>> wrote:
This message was sent securely using Zix® Hi Joe, We use ZapIT daily for routine CT QC and all weekly MRI QC. The process is super simple – we have preset protocols for the tech to run daily (CT) which include automatic forwarding to a
This message was sent securely using Zix®<https://urldefense.com/v3/__http://www.zixcorp.com/get-started/__;!!KGKeukY!wvC5uGQ8THHNh_vaRU_X7jFscVfD0Uh6P_RAFISE5Ue1nnSX2ySD_Gxv-_euqgr00Dej8M_8aQGuWtrngxUOnc70RBMxhSm9dgN-tKLD-No$>

Hi Joe,

We use ZapIT daily for routine CT QC and all weekly MRI QC.  The process is super simple – we have preset protocols for the tech to run daily (CT) which include automatic forwarding to a DICOM service which processes and then sends to ZapIT.  In about 2 minutes, the tech can open a ZapIT session, run the daily water/uniformity test, and the results are populated from the analysis program.  There is also full control to open and examine the images, with full DICOM headers and ROI/distance measurement tools for making manual verifications.  It is platform agnostic (we send both GE and Siemens test images), and because in NJ we have to keep the images for 30 days we have permanent retention in ZapIT.

Process is the same for MRI and the analysis covers the whole suite of phantom tests, including transmitter gain and CF.

This has really been a benefit for our staff in time saving and consistency; I actual gave a talk about our experiences with at AAPM 2 weeks ago and showed how it even exports ACR-format QC logs same as is on the ACR website.  And tech support is very fast, friendly and successful.

They do not, to my knowledge, have a nuc med solution at this time.  But we use almost exclusively GE cameras, and our daily floods are analyzed on-camera with quantitative results.  The staff just documents them in NMIS, so we don’t have a great need on the nuc side as it is fairly easy already (and the logging/reporting of long term QC results for the Discover-family scanners is not bad).

Dan

Daniel J. Januseski, M.S., DABR
Director, Department of Diagnostic Imaging Physics
Radiation Safety Officer
(856) 247-4436 | f (856) 247-2819
djanuseski at virtua.org<mailto:djanuseski at virtua.org>



From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu>> On Behalf Of Bob Kobistek via Intl_dxmedphys_wd_osu_list
Sent: Friday, August 15, 2025 12:36 PM
To: Meier, Joseph <Joseph.Meier at cchmc.org<mailto:Joseph.Meier at cchmc.org>>; DxMedPhys List <intl_dxmedphys_wd_osu_list at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>>
Cc: Brady, Samuel <Samuel.Brady at cchmc.org<mailto:Samuel.Brady at cchmc.org>>
Subject: [EXTERNAL] Re: [Intl_DxMedPhys] Seeking Daily CT and MR QC Automated Software

**WARNING** This email did NOT originate from inside Virtua.
Please STOP and ANALYZE before opening attachments, clicking on links, or providing any information.

Have you looked at Zap-It? Robert J. Kobistek, MS, FACR, DABR, MRSE(MRSC™) Medical Physicist RJK Medical Physics, Inc. 440-463-7879 From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces+bob=rjkmedphys.  com@  lists.  osu.  edu>
Have you looked at Zap-It? Robert J. Kobistek, MS, FACR, DABR, MRSE(MRSC™) Medical Physicist RJK Medical Physics, Inc. 440-463-7879 From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces+bob=rjkmedphys. com@ lists. osu. edu>
Have you looked at Zap-It?


Robert J. Kobistek, MS, FACR, DABR, MRSE(MRSC™)
Medical Physicist
RJK Medical Physics, Inc.
440-463-7879

From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces+bob=rjkmedphys.com at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list-bounces+bob=rjkmedphys.com at lists.osu.edu>> On Behalf Of Meier, Joseph via Intl_dxmedphys_wd_osu_list
Sent: Friday, August 15, 2025 11:32 AM
To: DxMedPhys List <intl_dxmedphys_wd_osu_list at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>>
Cc: Brady, Samuel <Samuel.Brady at cchmc.org<mailto:Samuel.Brady at cchmc.org>>
Subject: [Intl_DxMedPhys] Seeking Daily CT and MR QC Automated Software

Hello list serv, Do you know of any commercial software that does the following for Daily/Weekly QC: Automatically processes the: ACR MRI phantom Vendor specific CT QC phantoms. GE CT and Canon CT. Daily QC DICOM save screens from our GE Gamma
Hello list serv,

Do you know of any commercial software that does the following for Daily/Weekly QC:


  1.  Automatically processes the:

     *   ACR MRI phantom
     *   Vendor specific CT QC phantoms.  GE CT and Canon CT.
     *   Daily QC DICOM save screens from our GE Gamma and PET scanners.  It would be nice if the program could make direct measurements from the NM daily QC images.
     *   Can add other phantoms as needed.



  1.  Allows the technologists to log in, review the phantom scans for artifacts, and the quantitative results.   Afterwards they would indicate that the results have been reviewed or what actions were taken to address any issues. We would want them to be able to ww/wl, pan zoom etc.



  1.  Configurable email alerts.



  1.  Plots trends of the Daily QC analysis and perhaps predicts if maintenance will be needed… CT number drift, tranmit gain increases, etc.

Samuel Brady has developed Matlab code that processes all of our phantom scans on a chron job.  It sends out automated emails when there are issues with Daily QC.  This works very well for us, but we want to offload the upkeep of this program onto a commercial product.

Thanks for your suggestions,

-Joe Meier

Joseph Meier, PhD
Diagnostic Imaging Medical Physicist, Cincinnati Children’s Hospital
Assistant Professor, University of Cincinnati
3333 Burnet Ave, MLC 5031
Cincinnati, OH 45229-3026
(713) 422-3597 (O)
(513) 636-7775 (Administrative Assistant)


[cid:image003.png at 01DC10F6.F47638E0]



This email message is intended only for the use of the individual(s) or entity, to which it is addressed, and may contain information that is privileged, confidential and protected from disclosure under applicable law. If you are not the intended recipient, you are hereby notified that any dissemination, distribution or copying of this communication is strictly prohibited. If you have received this communication in error, please notify the sender immediately by telephone or return email, delete the message from your computer system, and return any hard copies to us at the following address 303 Lippincott Drive, Marlton, NJ 08053. Thank you.

This message was secured by Zix<https://urldefense.com/v3/__http://www.zixcorp.com__;!!KGKeukY!wvC5uGQ8THHNh_vaRU_X7jFscVfD0Uh6P_RAFISE5Ue1nnSX2ySD_Gxv-_euqgr00Dej8M_8aQGuWtrngxUOnc70RBMxhSm9dgN-EbpCmWs$>&#174.
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://lists.osu.edu/pipermail/intl_dxmedphys_wd_osu_list/attachments/20250819/fa544211/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image002.jpg
Type: image/jpeg
Size: 24070 bytes
Desc: image002.jpg
URL: <http://lists.osu.edu/pipermail/intl_dxmedphys_wd_osu_list/attachments/20250819/fa544211/attachment.jpg>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image003.png
Type: image/png
Size: 21970 bytes
Desc: image003.png
URL: <http://lists.osu.edu/pipermail/intl_dxmedphys_wd_osu_list/attachments/20250819/fa544211/attachment.png>


More information about the Intl_dxmedphys_wd_osu_list mailing list