[Intl_DxMedPhys] GE private DICOM tags for MRI
Steven Shea
steven.michael.shea at gmail.com
Thu Aug 14 18:31:43 EDT 2025
Wanted to add a quick note/helpful info that was passed to me at some point
in the past. DICOM conformance statements from GE (as well as other
vendors) will provide a lot of help when trying to find private tag
meanings or other system parameters. Here are links to GE & Siemens, but
all the vendors have these:
https://urldefense.com/v3/__https://www.gehealthcare.com/products/interoperability/dicom/magnetic-resonance-imaging-dicom-conformance-statements__;!!KGKeukY!yOAt5xr_a2PbQOzzLMbFEjf7N1DsPQkAMobeNZhegxI5-XawjmguzDCeqVyYavtIgjEslkg_yWfr2viPqdpqGhegDPeUi4nm3Ft9EfH9GdewC8fu$
https://urldefense.com/v3/__https://www.siemens-healthineers.com/services/it-standards/dicom-conformance-statements-magnetic-resonance__;!!KGKeukY!yOAt5xr_a2PbQOzzLMbFEjf7N1DsPQkAMobeNZhegxI5-XawjmguzDCeqVyYavtIgjEslkg_yWfr2viPqdpqGhegDPeUi4nm3Ft9EfH9GUaEOEuM$
You often have to wade through a lot of extraneous information, but it can
be very helpful for finding specific technical info (eg parallel imaging
On/Off, acceleration factor). Below are a mulitple GE DICOM tags that I
find very useful:
0018,0020: ScanningSequence
-
Ex: [EP\SE]
0018,0022: ScanOptions
-
Ex: [EPI_GEMS\FILTERED_GEMS\ACC_GEMS\PFF]
-
ACC_GEMS= Parallel image acceleration
-
EDR_GEMS= Extended Dynamic Range
0018,9087: DiffusionBValue (b value for particular image)
0019,105a: scan duration in us
0019,1093: Center Frequency (but not Hz, appears to be 1E8 = 10 MHz)
0019,1094: Transmit Gain
0019,1095: R1
0019,1096: R2
0019,109c: pulse sequence type
0043,102d: post-processing filter settings
-
P=PURE
-
S=SCIC
-
reFINE (PURE) and SCIC are features that improve image uniformity by
applying a pixel intensity correction. SCIC applies a unique
correction for
every image, which means that the same correctio43n will not be applied
even to images acquired at the same slice location. This may lead to
incorrect results when READY View or any post processing
software including
subtraction is used. As a result, it is strongly recommended to disable
SCIC for dynamic studies as well as any acquisition that generates more
than one image per slice location, for example, DWI or DTI. PURE can be
used as long as it uses the same PURE calibration series for all
the series
or images involved with the processing.
-
The reFINE (PURE) feature changes the pixel intensity in the image to
provide a more uniform image. This may lead to altered results when
CADstream is used in conjunction with PURE for processing MR images. A
modification to the CADstream difference threshold may be required. It is
recommended to either disable PURE for dynamic studies or
applying the PURE
feature uniformly across all the dynamic series.
-
s = SCENIC (?? Need to double-check)
-
A - Little sharpening, some smoothing: produces some smoothing while
sharpening a little to suppress noise. This filter provides some edge
enhancement of structural edges. Candidate images include 512×512 head,
abdomen, and extremity images.
-
B - High sharpening, some smoothing: a high sharpening filter with
little smoothing. This filter enhances structural edges. Used for high
signal and low noise 256×256 images. Candidate images include axial heads,
cervical spines, lumbar spines (small FOV), abdomens, and extremities.
-
C - Little sharpening, high smoothing: produces smoothing and very
little sharpening. This filter suppresses noise and provides smoother
structural edges. Candidate images include high-resolution and low-signal
images.
-
D - High sharpening, high smoothing: suppresses noise and provides
smoother regions and sharper structural edges. Used for high-resolution and
low-signal images.
-
E - MR angio filter: produces smoothing of vascular structures without
introducing black lines, and provides good vessel definition. Used for
source MIP1s from MRA2 studies.
-
F - Some sharpening, noise reduction : produces reduced image noise
while preserving edges. It also provides edge enhancement. Candidate images
include musculoskeletal and spine images.
-
G - Noise reduction: produces reduces image noise. Typically used with
neuro and spine images.
-
U: provides 5 levels (0-4) of noise reduction and sharpening options.
These can be individually adjusted and saved to suit your requirements.
Eg: P = PURE, D=smoothing, sharpening
0043,1039: b value for DWI sequences
-
Eg: [1000000800\8\0\0] --> b value of 800
0043,0x1081: GE Coil Name
-
Ex: [C-GE_HEAD]
0043,1083: Parallel Imaging Factors (Acceleration)
-
Eg: [0.5\1] --> 2x acceleration in phase-encode
0043,1084: Additional Parallel Imaging Data
-
Eg: [10000\11\-1\0\\YES\ARC]
-
Eg: [10000\0\-1\0\\NO\ ] --> no parallel imaging
0043,10B6: Multiband Parameters
0043,10CA: Recon DL Strength
-
Numerical representation like 0.30(/0.50/0.75)\ Literal representation
like LOW(/MEDIUM/HIGH)0.30\LOW, 0.50\Medium , or 0.75\High
Best regards, Steve
Steven Shea, PhD
Associate Professor, MR Scientist
Department of Radiology & Medical Imaging. Loyola Medicine
On Wed, Aug 6, 2025 at 11:01 AM <
intl_dxmedphys_wd_osu_list-request at lists.osu.edu> wrote:
> From: Bob Kobistek <bob at rjkmedphys.com>
> To: Bob Kobistek <bob at rjkmedphys.com>, International Diagnostic Medical
> Physics List <intl_dxmedphys_wd_osu_list at lists.osu.edu>
> Cc:
> Bcc:
> Date: Tue, 5 Aug 2025 21:57:53 +0000
> Subject: Re: [Intl_DxMedPhys] GE private DICOM tags for MRI
> Ignore this. 0018,0022. I must be getting old. It was right in front of
> me. I was looking for a private tag, but this is a public tag. Time for
> some vodka then a nap. Robert J. Kobistek, MS, FACR, DABR, MRSE(MRSC™)
> Medical Physicist RJK
>
> Ignore this. 0018,0022. I must be getting old. It was right in front of
> me. I was looking for a private tag, but this is a public tag.
>
>
>
> Time for some vodka then a nap.
>
>
>
>
>
> Robert J. Kobistek, MS, FACR, DABR, MRSE(MRSC™)
>
> Medical Physicist
>
> RJK Medical Physics, Inc.
>
> 440-463-7879
>
>
>
> *From:* Intl_dxmedphys_wd_osu_list
> <intl_dxmedphys_wd_osu_list-bounces+bob=rjkmedphys.com at lists.osu.edu> *On
> Behalf Of *Bob Kobistek via Intl_dxmedphys_wd_osu_list
> *Sent:* Tuesday, August 5, 2025 5:36 PM
> *To:* International Diagnostic Medical Physics List <
> intl_dxmedphys_wd_osu_list at lists.osu.edu>
> *Subject:* [Intl_DxMedPhys] GE private DICOM tags for MRI
>
>
>
> Hi folks I’m trying to fill out an ACR Phantom Data form for a GE MRI
> scanner. I neglected to record the scan options for the clinical scans.
> Does anybody know how to pull them from the private DICOM tags? Nothing
> jumps out at me, but
>
> Hi folks
>
>
>
> I’m trying to fill out an ACR Phantom Data form for a GE MRI scanner. I
> neglected to record the scan options for the clinical scans. Does anybody
> know how to pull them from the private DICOM tags? Nothing jumps out at me,
> but I know they have to be encoded in there somewhere because they display
> on the images when viewed on the AWS.
>
>
>
> Thanks.
>
>
>
>
>
> Robert J. Kobistek, MS, FACR, DABR, MRSE(MRSC™)
>
> Medical Physicist
>
> RJK Medical Physics, Inc.
>
> 440-463-7879
>
>
>
>
>
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://lists.osu.edu/pipermail/intl_dxmedphys_wd_osu_list/attachments/20250814/2a9813fc/attachment.html>
More information about the Intl_dxmedphys_wd_osu_list
mailing list