[Intl_DxMedPhys] CT Number Deviations on Siemens Force and Alpha – ACR Phantom Failures
Kalpana M. Kanal
kkanal at uw.edu
Thu Jul 24 10:21:42 EDT 2025
That's what was recommended to us as well for any measurements. We tested our Alpha last week and had no issues.
Kalpana
Kalpana M. Kanal, PhD, DABR, FACR, FAAPM, FSABI
Director, Diagnostic Physics Section and Resident Education
Program Director, Imaging Physics Residency Program
Professor, Radiology
Affiliate Faculty, Harborview Injury Prevention Center (HIPRC)
Adj. Professor, Oral Medicine
University of Washington | 1959 ne pacific street | box 357987 | seattle, wa 98195-7987 |
office location: 222F Portage Bay Bldg
email: kkanal at uw.edu<mailto:kkanal at uw.edu> | v: 206-616-9166 | pg: 206-680-6391 | f: 206-543-8356
________________________________
From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces+kkanal=uw.edu at lists.osu.edu> on behalf of Shane McCotter via Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu>
Sent: Thursday, July 24, 2025 7:17:34 AM
To: Imendra Ranatunge <ranatunge at gmail.com>
Cc: intl_dxmedphys_wd_osu_list at lists.osu.edu <intl_dxmedphys_wd_osu_list at lists.osu.edu>
Subject: Re: [Intl_DxMedPhys] CT Number Deviations on Siemens Force and Alpha – ACR Phantom Failures
Not sure if this will help, but try using the Quantative recon method (if I remember right). Had this issue with a Siemens before and apps recommended that recon. Shane Sent from my iPhone On Jul 24, 2025, at 8: 37 AM, Imendra Ranatunge via Intl_dxmedphys_wd_osu_list
Not sure if this will help, but try using the Quantative recon method (if I remember right). Had this issue with a Siemens before and apps recommended that recon.
Shane
Sent from my iPhone
On Jul 24, 2025, at 8:37 AM, Imendra Ranatunge via Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu> wrote:
Hi All, I’m encountering an issue with a couple of my Siemens CT scanners—specifically the Force and Alpha models. Several of my CT number measurements using the ACR phantom are failing for air, water, and bone. Some units fail all three, while
Hi All,
I’m encountering an issue with a couple of my Siemens CT scanners—specifically the Force and Alpha models. Several of my CT number measurements using the ACR phantom are failing for air, water, and bone. Some units fail all three, while others only one or two.
I tried placing a CTDI phantom adjacent to the ACR phantom, but it showed no improvement.
I reported the issue to Siemens. They performed a constancy test and stated that the units were functioning normally. However, my CT number results are still outside acceptable limits. In my experience, service engineers are often reluctant to perform time-consuming calibrations unless specifically requested.
Has anyone experienced this before? If so, do you know which specific calibration or adjustment Siemens needs to perform to correct these CT number inaccuracies?
Imendra