[Intl_DxMedPhys] Gamma Camera Max Count Rate Perpendicular versus Off-Angle

Januseski, Daniel DJANUSESKI at virtua.org
Tue Jun 24 12:34:15 EDT 2025


Agreed Nima – I also ignore since I perform the same geometry each year; the Intevo Bold comes in around 314-318 consistently between heads and YOY.  I’m using ~ 1 mCi.

Dan Januseski, MS, DABR
Voorhees, NJ



From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu> On Behalf Of Nima Kasraie via Intl_dxmedphys_wd_osu_list
Sent: Tuesday, June 24, 2025 11:14 AM
To: 'intl_dxmedphys_wd_osu_list at lists.osu.edu' <intl_dxmedphys_wd_osu_list at lists.osu.edu>
Subject: [EXTERNAL] Re: [Intl_DxMedPhys] Gamma Camera Max Count Rate Perpendicular versus Off-Angle

**WARNING** This email did NOT originate from inside Virtua.
Please STOP and ANALYZE before opening attachments, clicking on links, or providing any information.

I’ve also noticed angular dependencies. But I usually ignore, since I treat this as more of a consistency check. For my Symbia Intevo Bolds, I’ve set the pass/fail criteria at +/- 5% of Siemens reference value: One also notices variations depending
ZjQcmQRYFpfptBannerStart
This Message Is From an External Sender
This message came from outside your organization.
    Report Suspicious  <https://us-phishalarm-ewt.proofpoint.com/EWT/v1/A_-8kG5y!DkgK23pxIIIlTvUyYK-OqZBX5sudxxYejv2iknFKVScV-LHaytPwD0PjcYmKajcvmiZV6CjXnYWTfCZghJnmki9NTwKqpiyXcN7RIRPRH13v7dtWuX-W81zGJgSUYoq0bB0gi165hAhw$ >   ‌
ZjQcmQRYFpfptBannerEnd
I’ve also noticed angular dependencies. But I usually ignore, since I treat this as more of a consistency check. For my Symbia Intevo Bolds, I’ve set the pass/fail criteria at +/- 5% of Siemens reference value: One also notices variations depending
I’ve also noticed angular dependencies. But I usually ignore, since I treat this as more of a consistency check. For my Symbia Intevo Bolds, I’ve set the pass/fail criteria at +/- 5% of Siemens reference value:

[cid:image001.png at 01DBE504.4B318CD0]

One also notices variations depending on what window you use as well.


Plus, are you maybe too close to the head? I use 5mCi.


Nima

[cid:image002.png at 01DBE504.4B318CD0]
Nima Kasraie, PhD, MSc, DABR, DABSNM
Lead Physicist, Children’s Medical Center Dallas<https://urldefense.com/v3/__https:/www.childrens.com/__;!!KGKeukY!xmCo1aMNjkpqsHQud5hp1rl55EO3z0pSJokJKaTiOsB48aGqljvm-B-wiY7c6jw4L0ysTtlAYgMK017QAK_SV6UlD5joLP5_JsgMP246FAZd3KudalHipg$>
Associate Professor, Department of Radiology, UT Southwestern<https://urldefense.com/v3/__https:/profiles.utsouthwestern.edu/profile/187443/nima-kasraie.html__;!!KGKeukY!xmCo1aMNjkpqsHQud5hp1rl55EO3z0pSJokJKaTiOsB48aGqljvm-B-wiY7c6jw4L0ysTtlAYgMK017QAK_SV6UlD5joLP5_JsgMP246FAZd3Kuw8vs8nA$>
Phone: 214-648-7978
[cid:image003.png at 01DBE504.4B318CD0]

From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces+nima.kasraie=utsouthwestern.edu at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list-bounces+nima.kasraie=utsouthwestern.edu at lists.osu.edu>> On Behalf Of Dylan DeAngelis via Intl_dxmedphys_wd_osu_list
Sent: Tuesday, June 24, 2025 9:41 AM
To: 'intl_dxmedphys_wd_osu_list at lists.osu.edu' <intl_dxmedphys_wd_osu_list at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>>
Subject: [Intl_DxMedPhys] Gamma Camera Max Count Rate Perpendicular versus Off-Angle

Good morning, We assess max count rate on a gamma camera by moving a point source (usually 0.  5-1 mCi) closer to the detector until we observe the peak rate, before it starts to fall off again due to the paralyzable dead time. I’ve
Good morning, We assess max count rate on a gamma camera by moving a point source (usually 0. 5-1 mCi) closer to the detector until we observe the peak rate, before it starts to fall off again due to the paralyzable dead time. I’ve always
Good morning,

We assess max count rate on a gamma camera by moving a point source (usually 0.5-1 mCi) closer to the detector until we observe the peak rate, before it starts to fall off again due to the paralyzable dead time. I’ve always done this with the activity perpendicular to the surface of the detector. But we recently noticed that if you approach the detector at an angle, you can reach a higher count rate. I don’t know if this is true for every unit – I noticed it first on a Siemens Symbia and then made the same observation on an old Philips Vertex. The change is pretty substantial. On the Vertex, measuring perpendicular I observed around 205 kcps, but from an angle I managed to get 236 kcps.

In both cases, the dead time of the counting system should be fixed, so there should only be one max count rate, right? How could I be getting two different values based on the angle of approach? Has anyone else seen this?

Dylan DeAngelis, MS, DABR
NYS Licensed Medical Physicist
(Diagnostic Radiological Physics and Medical Nuclear Physics)

Upstate Medical Physics – Diagnostic Radiology, Medical Nuclear & Medical Health Physics, P.C.
W: 585-924-0350
C: 585-441-5096
https://urldefense.com/v3/__http://www.upstatemp.com__;!!KGKeukY!1yqkMMCo4EZ4VLT14XMYapz87KnOuCKatdeShovZO4gZvL3xD9cMqGCj37awEhNHh8oTMQfG1q-Rd07Eq_ZvyI-bbSNB9as-qME56sKttSI$ <https://urldefense.com/v3/__http:/www.upstatemp.com/__;!!KGKeukY!39QZ7SJnDoJxMrQB58D_ownz5P5PUFQZ3uNp484K1G8ywxH2def6_rUp_NxJOvwoJbC3hOIEKZT4upv0XrQI80ZFOTyj772nwdeesyXh8aDo$>

Click here<https://urldefense.com/v3/__https:/ws.onehub.com/signin__;!!KGKeukY!39QZ7SJnDoJxMrQB58D_ownz5P5PUFQZ3uNp484K1G8ywxH2def6_rUp_NxJOvwoJbC3hOIEKZT4upv0XrQI80ZFOTyj772nwdees-C9COca$> for Credentials (Login & Password required)
Click here<https://urldefense.com/v3/__https:/www.upstatemp.com/calibration-records/__;!!KGKeukY!39QZ7SJnDoJxMrQB58D_ownz5P5PUFQZ3uNp484K1G8ywxH2def6_rUp_NxJOvwoJbC3hOIEKZT4upv0XrQI80ZFOTyj772nwdeesweUt2pV$> for Equipment Calibration Records


________________________________

UT Southwestern

Medical Center

The future of medicine, today.



This message, and any included attachments, are from Virtua Health or its related affiliates and is intended only for the addressee(s). The information contained herein is privileged, proprietary or may include confidential information and/or protected patient health information. Any unauthorized review, forwarding, printing, copying, distributing, or otherwise disseminating or taking any action based on such information is strictly prohibited. If you have received this message in error, or have reason to believe you are not authorized to receive it, please delete this message promptly and notify the sender by e-mail with a copy to ISSECURITY at virtua.org. 

Thank you



This message was secured by Zix(R).
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20250624/2cf51b54/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image001.png
Type: image/png
Size: 17002 bytes
Desc: image001.png
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20250624/2cf51b54/attachment.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image002.png
Type: image/png
Size: 10985 bytes
Desc: image002.png
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20250624/2cf51b54/attachment-0001.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image003.png
Type: image/png
Size: 31503 bytes
Desc: image003.png
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20250624/2cf51b54/attachment-0002.png>


More information about the Intl_dxmedphys_wd_osu_list mailing list