[Intl_DxMedPhys] [External] Re: The end of LNT and ALARA?

Douglas Pfeiffer xraydoug at me.com
Thu May 29 12:55:33 EDT 2025


Also, ICRP TG91 (Radiation Risk Inference at Low-dose and Low-dose Rate Exposure for Radiological Protection Purposes) has a new report out for review. You can watch a workshop and download a copy of the draft at https://urldefense.com/v3/__https://www.icrp.org/page.asp?id=701__;!!KGKeukY!0YsogHwjx7vdMOi3n8LIl63gFYUhD7EeRTbmFxNpdh7a8srQ1e4CIBcYxeEnBgK2_fauTr5bcuEPuO6kYlCXw7XCKYLeHa6jFE9Y$ .

Best regards,
Doug



Douglas Pfeiffer, MS, DABR FACR, FAAPM
Medical Physicist, Radiation Safety Officer
Boulder Community Health
xraydoug at me.com
303.415.7515

+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-+-

The opinions expressed in this message are the product of the gray and white matter loitering in my cranium. I speak for myself and no one else, unless I say otherwise.

> On May 29, 2025, at 10:49 AM, Gress, Dustin via Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu> wrote:
> 
> NCRP has weighed in: 
>  
> NCRP Commentary No. 27 (2018) “provides a review of recent epidemiologic studies of low linear-energy transfer radiation, primarily at low doses or low dose rates, and an evaluation of whether the new observations provide support for the LNT model as used in radiation protection today.” I’ll let everyone look up the NCRP’s conclusions on their own, thanks to the excellent benefit of NCRP publication access provided by AAPM to its members.
>  
> Dustin
>  
> Dustin A. Gress, MS, FACR, FAAPM
> Senior Advisor for Medical Physics
> American College of Radiology
> *posting on behalf of myself and not my employer*
> dgress at acr.org <mailto:dgress at acr.org> | 703.648.2905
> LinkedIn <https://urldefense.com/v3/__http://www.linkedin.com/in/dustin-gress__;!!KGKeukY!wNvc8q-7xGIRk-VmgmQczW2DwwwtbxkV_rWSOAeDWayFbg2rzmz2ckOVQlEOPCGtYtvcx1P4x38Uy14p1Fq5BH9GNXAz6c1L1IA$> | twitter: @DustinGress <https://urldefense.com/v3/__https://x.com/DustinGress__;!!KGKeukY!wNvc8q-7xGIRk-VmgmQczW2DwwwtbxkV_rWSOAeDWayFbg2rzmz2ckOVQlEOPCGtYtvcx1P4x38Uy14p1Fq5BH9GNXAzI_pIYZ8$>
>  
>  
> From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu <mailto:intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu>> On Behalf Of Bob Pizzutiello via Intl_dxmedphys_wd_osu_list
> Sent: Thursday, May 29, 2025 12:43 PM
> To: Nima Kasraie <Nima.Kasraie at utsouthwestern.edu <mailto:Nima.Kasraie at utsouthwestern.edu>>
> Cc: DxMedPhys List <intl_dxmedphys_wd_osu_list at lists.osu.edu <mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>>
> Subject: [External] Re: [Intl_DxMedPhys] The end of LNT and ALARA?
>  
> CAUTION: This email originated from outside of the organization. Do not click links or open attachments unless you recognize the sender and know the content is safe.
>  
> Not an unreasonable expectation. My impression, based on co-authoring a single report published in the last decade. The reality is that while NCRP was chartered by Congress, funding is another matter and always a challenge. And their process
> Not an unreasonable expectation.  
>  
> My impression, based on co-authoring a single report published in the last decade. 
>  
> The reality is that while NCRP was chartered by Congress, funding is another matter and always a challenge. And their process is slow. 
>  
> After deciding on whether to accept a new project, it would be entered into the queue, then experts convened, meetings, numerous drafts, etc. Even if there were funding and they were willing to accept this project in the current political environment, a report would be years in the future. 
>  
> Perhaps others with closer connections to NCRP could correct or clarify my impression. I could be all wet. 
>  
> Bob
> From Bob Pizzutiello, typos courtesy iPhone
> 
> 
> On May 29, 2025, at 12:28 PM, Nima Kasraie via Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu <mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>> wrote:
> 
> 
> I would hope that’s where organizations such as the NCRP play an essential role? If I recall, the NCRP was specifically chartered by Congress in 1964 to develop and disseminate recommendations for radiation protection, essentially serving as
> I would hope that’s where organizations such as the NCRP play an essential role? If I recall, the NCRP was specifically chartered by Congress in 1964 to develop and disseminate recommendations for radiation protection <https://urldefense.com/v3/__https:/www.congress.gov/88/statute/STATUTE-78/STATUTE-78-Pg320.pdf__;!!KGKeukY!y44m5ZMMgS4Z34uxm6AHYoPUUX1xDXoO2EcvASumt-m9WZbM5aAwOuGyZ0NrPeEys8HoiBVqxKZEPV6uNZCfEEXJtlOpXP8lj4F90X97Ygff9mJ9bvBB2A$>, essentially serving as the source for which the NRC/EPA/DOE/FDA draw their expertise from to develop or revise regulations, guidelines, and operational procedures.
>  
> Nima 
>  
> <image001.png>
> Nima Kasraie, PhD, MSc, DABR, DABSNM
> Lead Physicist, Children’s Medical Center Dallas <https://urldefense.com/v3/__https:/www.childrens.com/__;!!KGKeukY!y44m5ZMMgS4Z34uxm6AHYoPUUX1xDXoO2EcvASumt-m9WZbM5aAwOuGyZ0NrPeEys8HoiBVqxKZEPV6uNZCfEEXJtlOpXP8lj4F90X97Ygff9mJzrMdtvg$>
> Associate Professor, Department of Radiology, UT Southwestern <https://urldefense.com/v3/__https:/profiles.utsouthwestern.edu/profile/187443/nima-kasraie.html__;!!KGKeukY!y44m5ZMMgS4Z34uxm6AHYoPUUX1xDXoO2EcvASumt-m9WZbM5aAwOuGyZ0NrPeEys8HoiBVqxKZEPV6uNZCfEEXJtlOpXP8lj4F90X97Ygff9mILQfgdZw$>
> Phone: 214-648-7978
> <image002.png>
>  
> From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu <mailto:intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu>> On Behalf Of Garrido, Mauricio via Intl_dxmedphys_wd_osu_list
> Sent: Tuesday, May 27, 2025 2:27 PM
> To: DxMedPhys List <intl_dxmedphys_wd_osu_list at lists.osu.edu <mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>>; Mary Ellen Jafari <MaryEllen.Jafari at kp.org <mailto:MaryEllen.Jafari at kp.org>>; Robert Morrison <robpmorrison at gmail.com <mailto:robpmorrison at gmail.com>>
> Subject: Re: [Intl_DxMedPhys] The end of LNT and ALARA?
>  
> It does seem like the DOE, DOD, and EPA will be making those decisions, most likely without medical physicists at the table. However, can we work with the Joint Commission or some other organization to at least bring that discussion to healthcare?
> It does seem like the DOE, DOD, and EPA will be making those decisions, most likely without medical physicists at the table. However, can we work with the Joint Commission or some other organization to at least bring that discussion to healthcare?
> It does seem like the DOE, DOD, and EPA will be making those decisions, most likely without medical physicists at the table. However, can we work with the Joint Commission or some other organization to at least bring that discussion to healthcare? The government is going to do whatever it's going to do to rapidly build nuclear reactors, but can we influence healthcare, so that we can keep ALARA and maybe work on LNT at this level?
>  
>  
> Mauricio Garrido, PhD
> Imaging Physics Fellow
> Department of Imaging Physics
> The University of Texas MD Anderson Cancer Center
> 1400 Pressler St, Houston, TX 77030
> Cell: 713-838-6619
>  
>  
> From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu <mailto:intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu>> on behalf of Robert Morrison via Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu <mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>>
> Sent: Tuesday, May 27, 2025 1:46 PM
> To: DxMedPhys List <intl_dxmedphys_wd_osu_list at lists.osu.edu <mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>>; Mary Ellen Jafari <MaryEllen.Jafari at kp.org <mailto:MaryEllen.Jafari at kp.org>>
> Subject: [EXTERNAL] Re: [Intl_DxMedPhys] The end of LNT and ALARA? 
>  
> SLOW DOWN! - EXTERNAL SENDER: intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu <mailto:intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu>
> Be suspicious of tone, urgency, and formatting. Do not click/open links or attachments on a mobile device. Wait until you are at a computer to confirm you are absolutely certain it is a trusted source.
> If you are at all uncertain use the Report Phish button and our Cybersecurity team will investigate.
> 
> I wrote something up and sent it to the list but it did not get posted.  I am not sure if it was because of something I did not do right or because someone did not want to post it because it looked to political which I was not trying to make it that way and stated that specifically in the post.  But one of my questions was the same as yours Mary Ellen, if replacing the existing standard way of setting limits is going to happen (which that is pretty clear from the EO) who will be the rule/decision makers for that new directive.  I understand that people here from this list like yourself would want to have a seat at the table.  But from the EO I read the DOE, DOD and EPA would be the ones making that decision.
>  
> This would be more for the public who live and work around these new Nuclear Reactors I would assume and of course for those who work at them.
>  
> I found it interesting that Europe seems to have stricter limits for the public as well as radiation workers than we do in the US.  I learned that a German airline actually has measured the radiation levels for all of their air routes and they control the staff to limit their exposure based upon the routes that they fly.  Maybe all the airlines do that.  I do not know.
>  
> Rob Morrison
> From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces+robpmorrison=gmail.com at lists.osu.edu <mailto:intl_dxmedphys_wd_osu_list-bounces+robpmorrison=gmail.com at lists.osu.edu>> on behalf of Mary Ellen Jafari via Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu <mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>>
> Sent: Tuesday, May 27, 2025 14:20
> To: DxMedPhys List <intl_dxmedphys_wd_osu_list at lists.osu.edu <mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>>
> Subject: Re: [Intl_DxMedPhys] The end of LNT and ALARA?
>  
> I support replacing LNT with limits based on published research more applicable to medical and occupational exposures that that used by ICRP,  but I do not at all support removing ALARA. 
> 
> My main concern is who would be setting the new "determinate limits."  If that is going to happen, then what actions can we take to ensure physicists have a seat at that table so that a scientific, evidence-based methodology is used instead of an AHARA approach (As High as Recklessly Allowed) supported by vested interests?    This is somewhat like the new CMS mandate on CT Dose and Image Quality where the evidence used to support it had a clear conflict of interest.
> 
> Mary Ellen
> 
> Mary Ellen Jafari, MS, DABR (D,N), MRSE, MRSO, FACR, FAAPM, CIIP
> Chief Physicist and Regional Radiation Safety Officer
> 
> Kaiser Permanente
> Southern California Permanente Medical Group Medical Imaging Technology and Informatics
> 199 S. Los Robles Ave., Pasadena CA 91101
> 626-430-1231 (office mobile)
> Maryellen.Jafari at kp.org <mailto:Maryellen.Jafari at kp.org>
> 
> -----Original Message-----
> From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces+maryellen.jafari=kp.org at lists.osu.edu <mailto:intl_dxmedphys_wd_osu_list-bounces+maryellen.jafari=kp.org at lists.osu.edu>> On Behalf Of Gary via Intl_dxmedphys_wd_osu_list
> Sent: Tuesday, May 27, 2025 10:24 AM
> To: William Sensakovic via Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu <mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>>
> Subject: Re: [Intl_DxMedPhys] The end of LNT and ALARA?
> 
> Caution: This email came from outside Kaiser Permanente. Do not open attachments or click on links if you do not recognize the sender.
> 
> ______________________________________________________________________
> I hear what you're saying William.  Sounds like:
>         regulations being broken
>         environmental damage
> 
> There may be some of that.  If so, I lay the blame at the feet of the same regulators and policy makers getting pushed out now.  Did the last 30+ years of federal strategy and regulatory policy lead to nuclear power innovations that would make it safer and cheaper?  Did they give it the same incentives that solar and wind got?  I wonder where we would be today if they had. 
> 
> I also like to image physicists (lots of them) working on developing and maintaining advanced nuclear power generation methods, instead of boiling water.  We are still boiling water, for crying out loud.  That is just sad.  Can you image your car boiling water, instead of being gas or electric?  It really looks like we wasted those 30+ years, almost like the feds were trying to destroy instead of develop.  And we need power badly, to say nothing of how we are going to need it with the rise of AI and other new tech.  It seems unthinkable that we would just keep burning fossil fuels.  Does anybody want that?
> 
> 
> ---
> Thanks,
> Gary Isenhower
> garyi at paxradia.com <mailto:garyi at paxradia.com>
> 
> 
> On Sun, 25 May 2025 19:43:20 +0000, William Sensakovic via Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu <mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>> wrote:
> 
> >I completely agree with Gary that this is the time to invest. The more
> >the relevant regulatory agencies are gutted, the less the short term
> >costs for spinning up new sites. Further, reduction in oversight costs
> >will be lower due to less regulations and fewer people to record
> >regulations being broken and fewer of the existing regulations being
> >enforced. Not to mention the reduced cost of keeping safety people on
> >staff. This results in lower medium range costs as well. Long term
> >costs.... well it depends what the cost share is for environmental
> >damage and fallout from nuclear accidents. If the government takes on
> >the burden, then there will be relatively little cost to the industry
> >except public trust, but that might not be much of a problem with
> >rebranding and clever marketing. Also they are unlikely to develop for
> >a decade or so. Hard to tell.
> >
> >I will let everyone determine for themselves whether recent changes to
> >the regulatory and emergency response landscape set off any alarm bells
> >or constitute best practice when reading the lessons learned from
> >Fukushima by the NRC
> >https://urldefense.com/v3/__https://www.ncbi.nlm.nih.gov/books/NBK25392 <https://urldefense.com/v3/__https:/www.ncbi.nlm.nih.gov/books/NBK25392>
> >3/__;!!KGKeukY!yy19_Yyu4uIBTivCLHEBRsaDDRSRNs_-OR-XqL5MbNhMFoy6aA8cPquD
> >-8DzSmaD8jmgvhUJhPJDGgBeMdkxb09ouzBmNQ0HHQ3jpt0u$
> >Or how political involvement, regulations, and organized crime
> >intertwine in the right environment
> >https://urldefense.com/v3/__https://www.scientificamerican.com/article/ <https://urldefense.com/v3/__https:/www.scientificamerican.com/article/>
> >special-report-help-wanted-in-fukushima-crime/__;!!KGKeukY!yy19_Yyu4uIB
> >TivCLHEBRsaDDRSRNs_-OR-XqL5MbNhMFoy6aA8cPquD-8DzSmaD8jmgvhUJhPJDGgBeMdk
> >xb09ouzBmNQ0HHTZt3jYp$ I don't want this to get "political" so I think
> >I will suggest the mantra of "do your own research" I think the cited
> >articles are stick to the science and are a good start.
> >
> >William F. Sensakovic, PhD, MRSC(MRSE), FAAPM Chair and Assoc. Prof.,
> >Mayo Clinic (Arizona) Founder, Telerad Physics Teaching, LLC
> >Twitter: @wfsensak
> >________________________________
> >From: Intl_dxmedphys_wd_osu_list
> ><intl_dxmedphys_wd_osu_list-bounces+wfsensak=outlook.com at lists.osu.edu <mailto:intl_dxmedphys_wd_osu_list-bounces+wfsensak=outlook.com at lists.osu.edu>>
> >on behalf of Gary via Intl_dxmedphys_wd_osu_list
> ><intl_dxmedphys_wd_osu_list at lists.osu.edu <mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>> Sent: Saturday, May 24,
> >2025 8:17 PM To: intl_dxmedphys_wd_osu_list at lists.osu.edu <mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>
> ><intl_dxmedphys_wd_osu_list at lists.osu.edu <mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>> Subject: Re:
> >[Intl_DxMedPhys] The end of LNT and ALARA?
> >
> >One thing it may mean is a rebirth of nuclear energy for the US.  Pro
> >tip: now is the time to invest.
> >
> >
> >---
> >Thanks,
> >Gary Isenhower
> >
> >
> >On Sat, 24 May 2025 17:24:33 +0000, Mary Ellen Jafari via
> >Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu <mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>>
> >wrote:
> >
> >>I didn't post the announcement from the White House to generate
> >>political commentary but rather to start a discussion on implications
> >>of such a change to radiology, medical physics, and occupational
> >>radiational workers, and how our professional organizations should
> >>respond to the Executive Order.
> >>
> >>I'm not the list moderator but is it possible to keep the discussion
> >>focused on those issues?
> >>
> >>Thank you.
> >>Mary Ellen
> >>
> >>Mary Ellen Jafari, MS, DABR (D,N), MRSE, MRSO, FACR, FAAPM, CIIP Chief
> >>Physicist and Regional Radiation Safety Officer
> >>
> >>Kaiser Permanente
> >>Southern California Permanente Medical Group Medical Imaging
> >>Technology and Informatics 199 S. Los Robles Ave., Pasadena CA 91101
> >>626-430-1231 (office mobile)
> >>Maryellen.Jafari at kp.org <mailto:Maryellen.Jafari at kp.org>
> >>
> >>-----Original Message-----
> >>From: Intl_dxmedphys_wd_osu_list
> >><intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu <mailto:intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu>> On Behalf Of Nicole
> >>Ranger via Intl_dxmedphys_wd_osu_list Sent: Saturday, May 24, 2025
> >>10:19 AM To: David Clunie <dclunie at dclunie.com <mailto:dclunie at dclunie.com>> Cc:
> >>intl_dxmedphys_wd_osu_list at lists.osu.edu <mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu> Subject: Re: [Intl_DxMedPhys]
> >>The end of LNT and ALARA?
> >>
> >>Caution: This email came from outside Kaiser Permanente. Do not open
> >>attachments or click on links if you do not recognize the sender.
> >>
> >>______________________________________________________________________
> >>I can’t help but feel that all these maneuvers are distractions. Also
> >>would be interested to learn of any binding agreements between the
> >>U.S. and the IAEA that are predicated on the long-standing and stable
> >>agreements around these regulatory concepts and limits.  International
> >>shipping of radioactive sources, enforcement of waste handling
> >>procedures, etc…
> >>
> >>These concepts should be agreed upon through scientific consensus and
> >>not political grandstanding.
> >>
> >>Nicole
> >>
> >>
> >>Sent from my iPhone
> >> 
> >>> On May 24, 2025, at 12:09 PM, David Clunie via
> >>> Intl_dxmedphys_wd_osu_list
> >>> <intl_dxmedphys_wd_osu_list at lists.osu.edu <mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>> wrote:
> >>>
> >>> Maybe this is a great way to improve genetic diversity in the US
> >>> population with more mutations, to substitute for a lack of
> >>> immigration.
> >>
> >>NOTICE TO RECIPIENT:  If you are not the intended recipient of this
> >>e-mail, you are prohibited from sharing, copying, or otherwise using
> >>or disclosing its contents.  If you have received this e-mail in
> >>error, please notify the sender immediately by reply e-mail and
> >>permanently delete this e-mail and any attachments without reading,
> >>forwarding or saving them. v.173.295  Thank you.
> >
>  
> The information contained in this e-mail message may be privileged, confidential, and/or protected from disclosure. This e-mail message may contain protected health information (PHI); dissemination of PHI should comply with applicable federal and state laws. If you are not the intended recipient, or an authorized representative of the intended recipient, any further review, disclosure, use, dissemination, distribution, or copying of this message or any attachment (or the information contained therein) is strictly prohibited. If you think that you have received this e-mail message in error, please notify the sender by return e-mail and delete all references to it and its contents from your systems.
>  
> UT Southwestern
> 
> Medical Center
> 
> The future of medicine, today.
> 
> 
> ______________________________________________________________________
> This email has been scanned by the Symantec Email Security.cloud service.
> ______________________________________________________________________
> 
> This message (including any attachments) is intended only for the use of the individual or entity to which it is addressed. It may contain information that is non-public, proprietary, privileged, confidential, and exempt from disclosure under applicable law and/or may constitute attorney work product. If you are not the intended recipient, do not use, disseminate, distribute, or copy this communication. If you have received this communication in error, please notify us immediately by telephone and destroy this message if this is a facsimile or (ii) delete this message immediately if this is an electronic communication.
> 
> 
> ______________________________________________________________________
> This email has been scanned by the Symantec Email Security.cloud service.
> ______________________________________________________________________

-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20250529/fbd817eb/attachment.html>


More information about the Intl_dxmedphys_wd_osu_list mailing list