Our experience has been about the same – called them out on failing collimation; told that it only has to pass 1 set (square) collimation test.  Now they acknowledge it and sent the product warning to do the upgrade.

Been chasing them for several months…no timeframe to do the upgrade yet…super frustrating



Dan Januseski, MS, DABR
Voorhees, NJ



From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces+djanuseski=virtua.org at lists.osu.edu> On Behalf Of Rebecca Milman via Intl_dxmedphys_wd_osu_list
Sent: Tuesday, May 20, 2025 6:26 PM
To: Goode, Allen R. *HS <ARG2N at uvahealth.org>
Cc: intl_dxmedphys_wd_osu_list at lists.osu.edu
Subject: [EXTERNAL] Re: [Intl_DxMedPhys] OEC Elite field size issue

**WARNING** This email did NOT originate from inside Virtua.
Please STOP and ANALYZE before opening attachments, clicking on links, or providing any information.

Allen, Thanks for posting this. I complained to OEC a couple of years ago when we got several of these units. They tried to deny there was an issue, saying that if it met the field restriction requirements in the diagonal directions then it
ZjQcmQRYFpfptBannerStart
This Message Is From an External Sender
This message came from outside your organization.
    Report Suspicious  <https://us-phishalarm-ewt.proofpoint.com/EWT/v1/A_-8kG5y!DMgKOvRdQMyvT3fSaAHhSunGYpsoWKE-FfDZ_xhWgbQ8NxwERAFgc6pV3d2NSjXMkkuOrrFz80d7pCA2h_hUIqK_Xau4C8mPhaGI90ZhAUB4MFviva-51NsWjJKj$ >   ‌
ZjQcmQRYFpfptBannerEnd
Allen, Thanks for posting this. I complained to OEC a couple of years ago when we got several of these units. They tried to deny there was an issue, saying that if it met the field restriction requirements in the diagonal directions then it
Allen,

Thanks for posting this. I complained to OEC a couple of years ago when we got several of these units. They tried to deny there was an issue, saying that if it met the field restriction requirements in the diagonal directions then it was okay. They then questioned my measurement methodology. Finally, they told the hospital that they (OEC) had to send someone out to do extra work to meet the requirements of "their physicist" (i.e., me). I sent an inquiry to the FDA at the time - who verified that my interpretation of the regs was correct. I wouldn't call them out on this if they hadn't been so rude and patronizing about the whole thing.

I am honest enough to admit that I feel somewhat vindicated about the whole thing. The phrase "GE Healthcare has become aware of a non-conformance" also makes me laugh because they knew about this at least 2 years ago, and someone other than me likely pointed it out to them before that. Better late than never, I suppose!

Rebecca.

Rebecca Milman, Ph.D.
Division Chief, Radiological Sciences
Department of Radiology
University of Colorado Anschutz Medical Campus
Aurora, Colorado


On Tue, May 20, 2025 at 2:45 PM Goode, Allen R. *HS via Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>> wrote:
Listers- In the past there were discussions on the list about collimation/field size failures for OEC C-arms. Over the past year, our hospital had several collimation failures on new GE/OEC Elite C-arms, during acceptance testing. Yesterday
Listers-
In the past there were discussions on the list about collimation/field size failures for OEC C-arms. Over the past year, our hospital had several collimation failures on new GE/OEC Elite C-arms, during acceptance testing. Yesterday we received the attached product warning from GE/OEC on this very issue. Looks like GE/OEC will be replacing the “squircle” collimator on many units.

If you have OEC Elite C-arms, you may want to check if any are on the very long list to be replaced.
Allen

ALLEN R. GOODE, MS, DABR
Chief Diagnostic Medical Physicist
Department of Radiology & Medical Imaging
434.924.2163
434.531.2459 (cell)
434.982.3500 #2281 (pager)
[cid:image001.jpg at 01DBCA4A.FD7697C0]



This message, and any included attachments, are from Virtua Health or its related affiliates and is intended only for the addressee(s). The information contained herein is privileged, proprietary or may include confidential information and/or protected patient health information. Any unauthorized review, forwarding, printing, copying, distributing, or otherwise disseminating or taking any action based on such information is strictly prohibited. If you have received this message in error, or have reason to believe you are not authorized to receive it, please delete this message promptly and notify the sender by e-mail with a copy to ISSECURITY at virtua.org. 

Thank you



This message was secured by Zix(R).
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20250521/8ca21816/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image001.jpg
Type: image/jpeg
Size: 3736 bytes
Desc: image001.jpg
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20250521/8ca21816/attachment.jpg>


More information about the Intl_dxmedphys_wd_osu_list mailing list