[Intl_DxMedPhys] Agfa CR
Jamie Dormand
Jamie.Dormand at rmh.nhs.uk
Tue May 20 05:28:05 EDT 2025
Hi Jered,
We also decommissioned our last Agfa CR reader some years ago but looking at our old notes it appears that the SOP class (including DX for processing) can be selected in the properties of the DICOM store destination and/or other export destinations. This can be reached through the NX Servicing & configuration tool so you will likely need a service engineer to configure this on your behalf, which may not be much help for exporting any images in the immediate future I’m afraid!
Thanks,
Jamie Dormand
Deputy Head of Diagnostic Radiology Physics
The Royal Marsden NHS Foundation Trust
020 7808 2154
https://urldefense.com/v3/__http://www.royalmarsden.nhs.uk__;!!KGKeukY!xGi6fiweULn-KHmsL_Rklnx7Ve-qnDBS2IV3bItwNcxzFxUCyJze3thm3G8k7W9D3DTm9EFspZ7_22pY09e4AJlonLZBPfNIBsUGwIsLl-GWD2Z7Rw$ <https://urldefense.com/v3/__http://www.royalmarsden.nhs.uk/__;!!KGKeukY!xGi6fiweULn-KHmsL_Rklnx7Ve-qnDBS2IV3bItwNcxzFxUCyJze3thm3G8k7W9D3DTm9EFspZ7_22pY09e4AJlonLZBPfNIBsUGwIsLl-HMZZPY_Q$ >
[cid:image002.jpg at 01DBC970.A471C080]
From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu> On Behalf Of Mark Supanich via Intl_dxmedphys_wd_osu_list
Sent: Monday 19 May 2025 21:43
To: Jered Wells <wellsmedicalphysics at gmail.com>; Nicole Ranger <nicoletranger at yahoo.com>
Cc: intl_dxmedphys_wd_osu_list at lists.osu.edu
Subject: Re: [Intl_DxMedPhys] Agfa CR
Caution: "This message was sent from an external source. Please do not click links or open attachments unless you recognise the sender and know the content is safe"
That talk was from the early teens 😂 I don't think I've touched an AGFA CR system in the past 12 years or so, so unfortunately, I can't add any more insight. From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces+mark_supanich=rush. edu@ lists. osu. edu>
That talk was from the early teens 😂
I don't think I've touched an AGFA CR system in the past 12 years or so, so unfortunately, I can't add any more insight.
________________________________
From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces+mark_supanich=rush.edu at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list-bounces+mark_supanich=rush.edu at lists.osu.edu>> on behalf of Nicole Ranger via Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>>
Sent: Monday, May 19, 2025 6:44 AM
To: Jered Wells <wellsmedicalphysics at gmail.com<mailto:wellsmedicalphysics at gmail.com>>
Cc: intl_dxmedphys_wd_osu_list at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu> <intl_dxmedphys_wd_osu_list at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>>
Subject: Re: [Intl_DxMedPhys] Agfa CR
Rush Email Security
This email originated from outside of RUSH. Do not click links or attachments unless you recognize the sender and know that the content is safe. RUSH will never ask for user ID information via email.
Mark Supanich gave a talk where he described how to export “For Processing” images on an AGFA DR system. I presume it would be a similar process for CR. Perhaps he can chime in but his talk is appended in the interim. See slide 20 or so. 59-17259-54061-211
Mark Supanich gave a talk where he described how to export “For Processing” images on an AGFA DR system. I presume it would be a similar process for CR. Perhaps he can chime in but his talk is appended in the interim. See slide 20 or so.
[cid:image001.png at 01DBC96E.88657E60]
59-17259-54061-211<https://urldefense.com/v3/__https:/www.aapm.org/meetings/amos2/pdf/59-17259-54061-211.pdf__;!!KGKeukY!wB7JaxwYEuQaVBZbiYy4YwoGvkU6mEkVhu9IUQW4q09v1Nh-3_O32lp7Nj8seX6Yzlp66SzzBCyHnevakHOgfadoRfvhm3av_-gp2KdZJwImpA$>
PDF Document · 469 KB<https://urldefense.com/v3/__https:/www.aapm.org/meetings/amos2/pdf/59-17259-54061-211.pdf__;!!KGKeukY!wB7JaxwYEuQaVBZbiYy4YwoGvkU6mEkVhu9IUQW4q09v1Nh-3_O32lp7Nj8seX6Yzlp66SzzBCyHnevakHOgfadoRfvhm3av_-gp2KdZJwImpA$>
Nicole
Sent from my iPhone
On May 19, 2025, at 6:29 AM, Nicole Ranger <nicoletranger at yahoo.com<mailto:nicoletranger at yahoo.com>> wrote:
Sorry, I misspoke. I was describing the Konica Minolta interface. Apologies!
Sent from my iPhone
On May 19, 2025, at 6:26 AM, Nicole Ranger <nicoletranger at yahoo.com<mailto:nicoletranger at yahoo.com>> wrote:
Jared,
I’ve tested many similar systems. There are collections of acquisition types for QC testing in the User Interface with multiple options including Flat Field, Dark Image, QC Phantom, etc…. If my memory is correct, Test 1 is the option that best corresponds to “For Processing”. You basically build up a collection of acquisition types before taking exposures and then make sure you have selected the one of interest before initiating the exposure. This ensures the appropriate processing (or lack thereof) is applied.
I’ll see if I can find written documentation describing the different QC acquisition types.
Nicole
Sent from my iPhone
On May 19, 2025, at 1:26 AM, Jered Wells via Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>> wrote:
Hi List: I’m testing an x-ray system with an Agfa CR unit tomorrow and would like to know if anyone has a quick guide to obtaining FOR_PROCESSING images? Also, any tips on testing exposure index on Agfa CR are appreciated (e. g. formulation for
Hi List:
I’m testing an x-ray system with an Agfa CR unit tomorrow and would like to know if anyone has a quick guide to obtaining FOR_PROCESSING images? Also, any tips on testing exposure index on Agfa CR are appreciated (e.g. formulation for proprietary EI, correct image processing mode, calibration kVp /filter).
Best,
Jered Wells, PhD, DABR
https://urldefense.com/v3/__https://www.linkedin.com/in/jered-wells-phd-dabr/__;!!KGKeukY!xGi6fiweULn-KHmsL_Rklnx7Ve-qnDBS2IV3bItwNcxzFxUCyJze3thm3G8k7W9D3DTm9EFspZ7_22pY09e4AJlonLZBPfNIBsUGwIsLl-EiPqxyrQ$ <https://urldefense.com/v3/__https:/www.linkedin.com/in/jered-wells-phd-dabr/__;!!KGKeukY!0bhVYZK6PDibrJ3-CLPJJZpkTYTgQfPATzlLUVk3g7i3MzyeIod1afQMm6Jv9CWd7k7IcxK_wkClW0I0bzKi-CEgkU8aJVwxdZgGqMGCQqvg88oi$>
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20250520/26662473/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image001.png
Type: image/png
Size: 194079 bytes
Desc: image001.png
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20250520/26662473/attachment.png>