[Intl_DxMedPhys] Mammography Display QC

Niall Phelan Niall.Phelan at screeningservice.ie
Wed Apr 23 11:11:01 EDT 2025


Hi Dan,

Thanks for your response. I have just had a look at the ACR manual.
[cid:image004.png at 01DBB46A.4BBA1160]

We would easily achieve the <20% tolerance on the centre luminance value here.
However, the EU/UK guidelines specify <5% on the maximum luminance value which is much more difficult…and I can’t see where this constraint might have come from and whether it is appropriate.

The radiologists prefer the large format displays and they definitely provide for a neater workspace.

Niall

From: Januseski, Daniel <DJANUSESKI at virtua.org>
Sent: 23 April 2025 15:44
To: Niall Phelan <Niall.Phelan at screeningservice.ie>; intl_dxmedphys_wd_osu_list at lists.osu.edu
Subject: RE: Mammography Display QC

This message was sent securely using Zix®<https://urldefense.com/v3/__http://scanmail.trustwave.com/?c=8024&d=nPyI6B3JKbm9KdbB6jJaAUy1vfF9EDD1at1kWaDVYA&u=http*3a*2f*2fwww*2ezixcorp*2ecom*2fget-started*2f__;JSUlJSUlJQ!!KGKeukY!3Up9aU45klv9k6iPQ6DOXYnHxc3wW98ht-hTxV6qHO4Hgie6MQTGs8sxW67Vw2L3VVZAWzJJY3Zn6RtsGI0Rx7THe1YLhW1zt7iJYGIB005I58GZlSwGqQ$ >

Hello Niall,

Interesting question regarding the newer 12MP displays.  Here in the U.S., those of us following the ACR Digital Mammography accreditation program and using their QC guide have to test and maintain a 20% target accuracy at the center of paired displays.  So the requirements are 30% per display across the full display dimension, and +/- 20% between the 2 displays at the center.

In practice, I have not had any trouble meeting even a 5% matching tolerance on our displays; single-display performance is also consistently below 10% I think mostly due to uniformity correction algorithms built into the controllers.


Dan Januseski, MS, DABR
Voorhees, NJ



From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces+djanuseski=virtua.org at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list-bounces+djanuseski=virtua.org at lists.osu.edu>> On Behalf Of Niall Phelan via Intl_dxmedphys_wd_osu_list
Sent: Wednesday, April 23, 2025 4:09 AM
To: intl_dxmedphys_wd_osu_list at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>
Subject: [EXTERNAL] [Intl_DxMedPhys] Mammography Display QC

**WARNING** This email did NOT originate from inside Virtua.
Please STOP and ANALYZE before opening attachments, clicking on links, or providing any information.

On the subject of display QC….  . We have for many years employed a standard configuration of two 5MP displays side-by-side for mammography reporting and perform QC according to European guidelines (EFOMP & UK; relevant standards
ZjQcmQRYFpfptBannerStart
This Message Is From an External Sender
This message came from outside your organization.
    Report Suspicious  <https://urldefense.com/v3/__https://scanmail.trustwave.com/?c=8024&d=nPyI6B3JKbm9KdbB6jJaAUy1vfF9EDD1attsWqrSMQ&u=https*3a*2f*2fus-phishalarm-ewt*2eproofpoint*2ecom*2fEWT*2fv1*2fA*5f-8kG5y*21DugKe3uxIIIF7vUz4K4PJ7oq2WDt7jqw8CgW*5fWWxgvWN44Iu*5fbiCA1mS-kT8OUnnhVG3kmhnfiDx350Ynu*5f1EZhma*5fuutPc3vLsDihMxoDuzH13M4G47gmps85TGDhkX973fYcI7itaVPD-0oLE*24__;JSUlJSUlJSUlJSUlJSUl!!KGKeukY!3Up9aU45klv9k6iPQ6DOXYnHxc3wW98ht-hTxV6qHO4Hgie6MQTGs8sxW67Vw2L3VVZAWzJJY3Zn6RtsGI0Rx7THe1YLhW1zt7iJYGIB005I58HhMSKNBQ$ >   ‌
ZjQcmQRYFpfptBannerEnd
On the subject of display QC…. . We have for many years employed a standard configuration of two 5MP displays side-by-side for mammography reporting and perform QC according to European guidelines (EFOMP & UK; relevant standards extracted
On the subject of display QC…..

We have for many years employed a standard configuration of two 5MP displays side-by-side for mammography reporting and perform QC according to European guidelines (EFOMP & UK; relevant standards extracted in boxes below) which in turn are based on AAPM TG-18 and subsequently TG-270.
However, there is one additional constraint on the luminance variation between the two displays (<5%) which doesn’t appear to form part of the AAPM guidance. I have always felt this constraint makes sense for diagnostic reporting with two paired displays, and we have achieved this performance in practice for years (sometimes by adjusting the L(max) of both displays independently).
Equally, all the guidance documents specify the uniformity tolerance for a single display to be <30% but in practice, we typically measure this <10% across the entire luminance range.

Recently, while testing some new large format displays (12MP) which are split into two independent windows for side-by-side reporting, we haven’t been able to achieve this 5% tolerance on L(max) for what are effectively two sides of the same display. In one case, we measure >15% difference and of course the L(max) can’t be adjusted as before for two independent displays.
Equally, when we consider the large format display on its own or split into independent display windows, either configuration conforms to the 30% uniformity tolerance.

Ultimately, the questions are:
Does anyone in North America (or elsewhere) impose a similar constraint on paired displays?
Does it make sense to impose a 30% uniformity limit on each single display and then a tighter 5% constraint on the paired displays?
Is 30% too high anyway for luminance uniformity (CRTs could achieve this 20+ years ago)?


EFOMP:
[cid:image005.png at 01DBB46A.4BBA1160]

UK:
[cid:image006.png at 01DBB46A.4BBA1160]


Grateful to hear any thoughts on this,

Niall


Niall Phelan
Chief Physicist
BreastCheck | 36 Eccles Street | Dublin 7 | DO7 Y7C6

Tel: +353 1 8826247 | +353 87 6697095
Email: niall.phelan at screeningservice.ie<mailto:niall.phelan at screeningservice.ie> https://urldefense.com/v3/__http://www.screeningservice.ie__;!!KGKeukY!3Up9aU45klv9k6iPQ6DOXYnHxc3wW98ht-hTxV6qHO4Hgie6MQTGs8sxW67Vw2L3VVZAWzJJY3Zn6RtsGI0Rx7THe1YLhW1zt7iJYGIB005I58HLXg9v8A$ <https://urldefense.com/v3/__http://scanmail.trustwave.com/?c=8024&d=nPyI6B3JKbm9KdbB6jJaAUy1vfF9EDD1at5oDqPXNA&u=http*3a*2f*2fhttps*2a3a*2a2f*2a2fscanner*2a2etopsec*2a2ecom*2a2f*2a3ft*2a3dddab9df86276a2c67244b679c60f18671b56bf3e*2a26d*2a3d2120*2a26u*2a3dwww*2a2escreeningservice*2a2eie*2a26r*2a3dshow*5f*5f*3bJSUlJSUlJSUlJSUlJSUlJQ*21*21KGKeukY*213hrUJN-eI9nOaEqr5AAbwr7kQzhWHTYnATrEobbuFpG*5ffYv9NA*5fxty1vMVtYGIOVj2onapjctvJa5CkhQ3AX2IU1RKh6DGwNOgTscAeDv3DW4UOF2eyz4A*24__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJQ!!KGKeukY!3Up9aU45klv9k6iPQ6DOXYnHxc3wW98ht-hTxV6qHO4Hgie6MQTGs8sxW67Vw2L3VVZAWzJJY3Zn6RtsGI0Rx7THe1YLhW1zt7iJYGIB005I58FlJd--yA$ >
[cid:image007.jpg at 01DBB46A.4BBA1160]

________________________________

Scanned by MailMarshal - M86 Security's comprehensive email content security solution. Download a free evaluation of MailMarshal at https://urldefense.com/v3/__http://www.m86security.com__;!!KGKeukY!3Up9aU45klv9k6iPQ6DOXYnHxc3wW98ht-hTxV6qHO4Hgie6MQTGs8sxW67Vw2L3VVZAWzJJY3Zn6RtsGI0Rx7THe1YLhW1zt7iJYGIB005I58HbuAL1HA$ <https://urldefense.com/v3/__https://scanmail.trustwave.com/?c=8024&d=nPyI6B3JKbm9KdbB6jJaAUy1vfF9EDD1aok5CPHQMQ&u=https*3a*2f*2furldefense*2ecom*2fv3*2f*5f*5fhttp*3a*2fwww*2em86security*2ecom*5f*5f*3b*21*21KGKeukY*213hrUJN-eI9nOaEqr5AAbwr7kQzhWHTYnATrEobbuFpG*5ffYv9NA*5fxty1vMVtYGIOVj2onapjctvJa5CkhQ3AX2IU1RKh6DGwNOgTscAeDv3DW4UO5KzqMGQ*24__;JSUlJSUlJSUlJSUlJSUlJSUlJSUl!!KGKeukY!3Up9aU45klv9k6iPQ6DOXYnHxc3wW98ht-hTxV6qHO4Hgie6MQTGs8sxW67Vw2L3VVZAWzJJY3Zn6RtsGI0Rx7THe1YLhW1zt7iJYGIB005I58EjgWE0XA$ >
________________________________
This email and any files transmitted with it are confidential and intended solely for the individual or entity to whom they are addressed. If you received this email in error, please contact the sender immediately and delete the email and any information transmitted with it. If you are not the intended recipient of this email you should not disseminate, disclose, distribute, print, use, rely upon or copy this email or the contents of this email. Any views or opinions presented in this email are solely those of the author and do not necessarily represent those of the National Screening Service (NSS). As the security of this message cannot be guaranteed, the NSS does not accept any liability for any errors or omissions in the contents of this message arising as a result of email transmission. The recipient should check this email and any attachments for the presence of viruses. The NSS accepts no liability for any loss or damage caused by any virus transmitted by this email.

This email message is intended only for the use of the individual(s) or entity, to which it is addressed, and may contain information that is privileged, confidential and protected from disclosure under applicable law. If you are not the intended recipient, you are hereby notified that any dissemination, distribution or copying of this communication is strictly prohibited. If you have received this communication in error, please notify the sender immediately by telephone or return email, delete the message from your computer system, and return any hard copies to us at the following address 303 Lippincott Drive, Marlton, NJ 08053. Thank you.

This message was secured by Zix<https://urldefense.com/v3/__http://scanmail.trustwave.com/?c=8024&d=nPyI6B3JKbm9KdbB6jJaAUy1vfF9EDD1at1rCaqGbw&u=http*3a*2f*2fwww*2ezixcorp*2ecom__;JSUlJSU!!KGKeukY!3Up9aU45klv9k6iPQ6DOXYnHxc3wW98ht-hTxV6qHO4Hgie6MQTGs8sxW67Vw2L3VVZAWzJJY3Zn6RtsGI0Rx7THe1YLhW1zt7iJYGIB005I58GwnjIucw$ >®.

#####################################################################################
Scanned by MailMarshal - M86 Security's comprehensive email content security solution. 
Download a free evaluation of MailMarshal at https://urldefense.com/v3/__http://www.m86security.com__;!!KGKeukY!3Up9aU45klv9k6iPQ6DOXYnHxc3wW98ht-hTxV6qHO4Hgie6MQTGs8sxW67Vw2L3VVZAWzJJY3Zn6RtsGI0Rx7THe1YLhW1zt7iJYGIB005I58HbuAL1HA$ 
#####################################################################################

This email and any files transmitted with it are confidential and intended
solely for the individual or entity to whom they are addressed. If you received
this email in error, please contact the sender immediately and delete the
email and any information transmitted with it. If you are not the intended
recipient of this email you should not disseminate, disclose, distribute, print, 
use, rely upon or copy this email or the contents of this email. Any views or 
opinions presented in this email are solely those of the author and do not 
necessarily represent those of the National Screening Service (NSS). As the 
security of this message cannot be guaranteed, the NSS does not accept any 
liability for any errors or omissions in the contents of this message arising as 
a result of email transmission. The recipient should check this email and any 
attachments for the presence of viruses. The NSS accepts no liability for any 
loss or damage caused by any virus transmitted by this email.
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20250423/d4633f42/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image004.png
Type: image/png
Size: 21320 bytes
Desc: image004.png
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20250423/d4633f42/attachment.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image005.png
Type: image/png
Size: 15730 bytes
Desc: image005.png
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20250423/d4633f42/attachment-0001.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image006.png
Type: image/png
Size: 9791 bytes
Desc: image006.png
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20250423/d4633f42/attachment-0002.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image007.jpg
Type: image/jpeg
Size: 1345 bytes
Desc: image007.jpg
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20250423/d4633f42/attachment.jpg>


More information about the Intl_dxmedphys_wd_osu_list mailing list