Yes. We have also investigated global noise metrics and their potential
limitations in clinical applications. We are wrapping up our writing and
hope to share our results soon.

Thanks!

Jie

On Tue, Mar 11, 2025 at 11:32 AM Gauntt, David <davidgauntt at uabmc.edu>
wrote:

> Jie,
>
>
> Does your paper  address the issue of image quality?
>
>
> The new CMS measure addresses both dose and image quality, using "global
> noise" as a measure of image quality.  However, as near as I can tell the
> measure does not address how you measure "global noise" in a clinical
> image, and have not found any papers from the measure's stewards addressing
> this issue.
>
>
> Does anyone have any references to anything addressing how you calculate
> global noise?
>
>
> ---
> David M. Gauntt, Ph.D.
> Associate Professor
> Division of Physics and Engineering
> Department of Radiology
> (205)975-3777
>
> [image: 1647458080270]
>
>
>
>
>
> ------------------------------
> *From:* Intl_dxmedphys_wd_osu_list
> <intl_dxmedphys_wd_osu_list-bounces+davidgauntt=uabmc.edu at lists.osu.edu>
> on behalf of Jie Zhang via Intl_dxmedphys_wd_osu_list <
> intl_dxmedphys_wd_osu_list at lists.osu.edu>
> *Sent:* Tuesday, March 11, 2025 10:04 AM
> *To:* intl_dxmedphys_wd_osu_list at lists.osu.edu
> *Subject:* EXT MSG [Intl_DxMedPhys] Fwd: The New CMS Measure: Methods for
> Size-Adjusted Dose and Their Variabilities
>
> Thanks to Dr. Wunderle! I originally posted this to the old listserver and
> am now forwarding it to the new Dx list. I am sorry if you receive it
> twice. Jie ---------- Forwarded message --------- From: Jie Zhang <jzctjmn@
> gmail. com> Date:
> Thanks to Dr. Wunderle! I originally posted this to the old listserver and
> am now forwarding it to the new Dx list.  I am sorry if you receive it
> twice.
>
> Jie
>
> ---------- Forwarded message ---------
> From: Jie Zhang <jzctjmn at gmail.com>
> Date: Tue, Mar 11, 2025 at 9:15 AM
> Subject: The New CMS Measure: Methods for Size-Adjusted Dose and Their
> Variabilities
> To: DXIMGMEDPHYS at HERMES.GWU.EDU <dximgmedphys at hermes.gwu.edu>
>
>
> Dear Colleagues,
>
> The new CMS guidance on excessive radiation measurement allows flexibility
> in software selection, but this may lead to inconsistencies and undermine
> its goal.
>
> We recently evaluated five methods for estimating effective diameter and
> their impact on Size-Specific Dose Estimate (SAD) in thoracic and abdominal
> CT. A retrospective analysis of 719 exams revealed inconsistencies in SAD
> methods, dose thresholds, and application across patient populations and
> institutions. Addressing these issues is critical for accurate dose
> reporting and diagnostic integrity.
>
> Our manuscript is available on arXiv: https://urldefense.com/v3/__https://arxiv.org/abs/2503.06644__;!!KGKeukY!zmr_jN5LoiU_rleb5lpW2NzFy4K2O2n9yKh2SxlLia_7z3tw3DjqZAKsP0YXavEJMB2uDPouOMQTfjfJCT0Q1bkqMcvGZqringvsfg$ 
> <https://urldefense.com/v3/__https://arxiv.org/abs/2503.06644__;!!KGKeukY!wkvI7i3VRXaaElI73WdyPRf_Zk6qQqZ2CCRW0BoG-aYbTN1bVG_WIFbTeFOBYzuVEpUvax1OJJGkFDM4g68KPmxld3yXDnHKjwB6aQ$>
> .
>
> Our findings reinforce concerns raised by the AAPM Commissioned Panel,
> recently published in *AJR*, and we hope to provide a technical
> perspective on the CMS measure.
>
> Best Regards,
>
> Jie
>
> *Jie Zhang, PhD, DABR (D, N), FAAPM*
>
> Professor of Radiology and Biomedical Engineering
>
> Chief, Division of Diagnostic & Nuclear Medical Physics
>
> Program Director, Diagnostic Imaging Physics Residency
>
> Department of Radiology
>
> University of Kentucky College of Medicine
>
> UK HealthCare
>
> 800 Rose Street, Room HX-307
>
> Lexington, KY 40536-0293
>
> Phone: (859) 323-2954
>
> Email: jnzh222 at uky.edu
>
>>
>>
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20250311/ddef2dc1/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: OutlookEmoji-1647458080270d7f02ca5-59f0-4d92-8fab-6fcda46b3f08.png
Type: image/png
Size: 15009 bytes
Desc: not available
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20250311/ddef2dc1/attachment.png>


More information about the Intl_dxmedphys_wd_osu_list mailing list