Has anyone had to deal with these devices in the context of nuclear medicine? A few years ago, I was at a site where we implemented these policies at the behest of our quality department for PET/CT (due to CT, not PET, we specifically asked about this). At the time we did not have to do anything for general NM or more significantly RPT.
The paper linked below it states the following for PET, seems most vendors want it removed for the procedure and you can put it or a replacement back on when the procedure is done… but the patient is still radioactive.
[cid:image003.png at 01DB8CD7.D5F98B40]
I have never succeeded in finding a threshold dose rate, a cumulative dose, or a combination of both that would result in these devices not working in the context of x-ray, CT, fluoro, radiology, NM, PET, RPT, etc.
I can understand the vendor trying to save themselves from liability if a device fails after a radiology procedure. But having patients remove, turn off, replace, etc. these devices just because the vendor says “radiation may cause these devices to malfunction” just to prevent liability is doing our patients a huge disservice.
Thanks
Joe
From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces+joseph.steiner=bsd.uchicago.edu at lists.osu.edu> On Behalf Of McCann, Aaron via Intl_dxmedphys_wd_osu_list
Sent: Tuesday, March 4, 2025 5:43 AM
To: intl_dxmedphys_wd_osu_list at lists.osu.edu
Subject: [EXTERNAL] Re: [Intl_DxMedPhys] insulin pumps and CT
Hi all, Sites *not* employing an all-or-nothing policy should be aware that this is still a dynamic field and that further testing of these devices may result in changes to their formal safety status. For example, some (not all) FreeStyle Libre
ZjQcmQRYFpfptBannerStart
External: Use caution with links, attachments, and providing information.
Report Suspicious <https://us-phishalarm-ewt.proofpoint.com/EWT/v1/MyIu0v6UfBA57LoN!4d_Mzceq7tewe65GXrWTFcfhBryHFd9bqZNmzRHqbarp5nht54KwJIPooRgSE61ZCiQ4CrdkC1WKm9DUpOvdovyPdqp3uRd_VxxUa6zZxAvvFV07DxNOzB-J1Sqk3wx-aJqZh11wPyuHiwNTYIKSdA$>
ZjQcmQRYFpfptBannerEnd
Hi all, Sites *not* employing an all-or-nothing policy should be aware that this is still a dynamic field and that further testing of these devices may result in changes to their formal safety status. For example, some (not all) FreeStyle Libre
Hi all,
Sites *not* employing an all-or-nothing policy should be aware that this is still a dynamic field and that further testing of these devices may result in changes to their formal safety status. For example, some (not all) FreeStyle Libre device models have recently received FDA clearance as MR Conditional (and as safe to be worn during CT, x-ray and diathermy procedures), rendering the published table at the start of this thread out of date:
https://urldefense.com/v3/__https://www.medscape.com/viewarticle/newer-abbott-libre-cgm-sensor-use-now-okay-during-imaging-2024a1000k2f__;!!KGKeukY!w7L3G7cww6LP_CZ6O3KIgwuZva6STgiv7o-TXAORMxsjDsf_pR-ammkzXEzUykudGbMjnw5dct0lOjKS-L8Spw0qmrL0zN9rBUgo3G6NsQ2RSJpeIUeqpeE$ <https://urldefense.com/v3/__https:/www.medscape.com/viewarticle/newer-abbott-libre-cgm-sensor-use-now-okay-during-imaging-2024a1000k2f__;!!KGKeukY!1yQF-zi8pBgrZt7dRpfQlNwTOuUuyWkqZ7TVrQXQsopDp4yV6K4ld7Dgwx5Rf3vymWx8A8xw2ssfb4aVWHUi_xoOx656tZIZgXKO2cLIuNMqEfVRVrWz1K8$>
One other consideration: I am aware of at least one report where a patient was informed that their glucose monitor should be removed prior to MRI – they did so themselves a few hours before their examination, and subsequently suffered a hypoglycaemic episode during the scan. These devices are critical for some patients in managing their own conditions, and the duration without monitoring should be kept as short as possible. Having an advance plan in place for removal and replacement that involves the patient is important.
Kind regards,
Aaron
_____________________________
Dr Aaron McCann
Clinical Scientist
Radiological Imaging & Protection Service
Regional Medical Physics Service
Belfast Health & Social Care Trust
Forster Green Site
110 Saintfield Road
Belfast
Northern Ireland
BT8 6GR
Tel: 028 9504 4863
(Note: my normal office hours are Mon-Fri 8am – 4pm)
From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces+aaron.mccann=belfasttrust.hscni.net at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list-bounces+aaron.mccann=belfasttrust.hscni.net at lists.osu.edu>> On Behalf Of Szczykutowicz, Timothy P via Intl_dxmedphys_wd_osu_list
Sent: 03 March 2025 23:19
To: intl_dxmedphys_wd_osu_list at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>
Subject: Re: [Intl_DxMedPhys] insulin pumps and CT
The borders of the state are the borders of the university (badgers will get this reference) Attached is our current all or nothing policy. Thanks to all who have sent me their policies and links!!! I also just pinged my rad onc friends that
The borders of the state are the borders of the university (badgers will get this reference)
Attached is our current all or nothing policy.
Thanks to all who have sent me their policies and links!!!
I also just pinged my rad onc friends that live in the basement here. They brought up a good point…. What are you going to do for 25 fractions!?
-stick
Timothy P. Szczykutowicz, Ph.D., DABR
Associate Professor
Departments of Radiology, Medical Physics and BME
University of Wisconsin Madison
Cell# 1-716-560-7751
Office# 1-608-263-5729
he/him/his
From: Josh M Wilson <joshua.wilson at duke.edu<mailto:joshua.wilson at duke.edu>>
Sent: Monday, March 3, 2025 4:20 PM
To: Szczykutowicz, Timothy P <TSzczykutowicz at uwhealth.org<mailto:TSzczykutowicz at uwhealth.org>>; intl_dxmedphys_wd_osu_list at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>; Wunderle, Kevin <Kevin.Wunderle at osumc.edu<mailto:Kevin.Wunderle at osumc.edu>>
Subject: Re: insulin pumps and CT
You don't often get email from joshua.wilson at duke.edu<mailto:joshua.wilson at duke.edu>. Learn why this is important<https://urldefense.com/v3/__https:/aka.ms/LearnAboutSenderIdentification__;!!KGKeukY!0OUW5OkrU1mkaxwZm91wzoWYV97nYSEjtoVhCustFte0tslNQT5dN-D8oynIYskROSPNpsfefAMhQHEVE15bhBw_EQ5QRYv9gUPhSqcy6kZu1fGVYA$>
WARNING: This email appears to have originated outside of the UW Health email system.
DO NOT CLICK on links or attachments unless you recognize the sender and know the content is safe.
Deja vu; attached from the old listserv. As far as I can tell, no one publicly shared a policy to the last request. YMMV.
Best,
Josh M. Wilson, PhD
Radiation Physicist, Clinical Imaging Physics Group
Duke University Health System
Director, Duke Imaging Physics Residency Program
Assistant Professor, Department of Radiology
Duke University School of Medicine
________________________________
From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu>> on behalf of Wunderle, Kevin via Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>>
Sent: Monday, March 3, 2025 4:35 PM
To: Szczykutowicz, Timothy P <TSzczykutowicz at uwhealth.org<mailto:TSzczykutowicz at uwhealth.org>>; intl_dxmedphys_wd_osu_list at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu> <intl_dxmedphys_wd_osu_list at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>>
Subject: Re: [Intl_DxMedPhys] insulin pumps and CT
Hey Tim,
This is an important topic, and one that has substantial cost liability for patients who, for some units, must replace those devices, and, as you allude to, substantial liability for institutions that do anything other than what the device manufacturers state.
Perhaps this is a topic to bring to the AAPM MPI (Medical Physics Institute) for investigation as a potential project? Jatinder Palta is the director of this new endeavor and might be interested in hearing about this need.
Kind regards,
Kevin
[The Ohio State University]
Kevin Wunderle, PhD, FAAPM, FACR
Professor, Department of Radiology
Diagnostic Medical Physicist
395 W. 12th Avenue,
Floor 4, Room 424
Columbus, OH, 43210
216-245-5513 Mobile
kevin.wunderle at osumc.edu<mailto:kevin.wunderle at osumc.edu>
From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces+kevin.wunderle=osumc.edu at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list-bounces+kevin.wunderle=osumc.edu at lists.osu.edu>> On Behalf Of Szczykutowicz, Timothy P via Intl_dxmedphys_wd_osu_list
Sent: Monday, March 3, 2025 4:24 PM
To: intl_dxmedphys_wd_osu_list at lists.osu.edu<mailto:intl_dxmedphys_wd_osu_list at lists.osu.edu>
Subject: [Intl_DxMedPhys] insulin pumps and CT
Anyone have a policy for insulin pumps and sensors other than “taken them off” for CT they are willing to share? huge risk if they malfunction…possible death, so I get why folks defer to the vendor policies which don’t
Anyone have a policy for insulin pumps and sensors other than “taken them off” for CT they are willing to share?
huge risk if they malfunction…possible death, so I get why folks defer to the vendor policies which don’t appear to be actually tested inside a scanner.
I found this paper, link below. But it basically says listen to the vendors. It seems in practice, this would be complicated if we have to have a workflow where each device is looked up.
https://urldefense.com/v3/__https://link.springer.com/article/10.1007/s11892-021-01375-7__;!!KGKeukY!w7L3G7cww6LP_CZ6O3KIgwuZva6STgiv7o-TXAORMxsjDsf_pR-ammkzXEzUykudGbMjnw5dct0lOjKS-L8Spw0qmrL0zN9rBUgo3G6NsQ2RSJpe_Z755q0$ <https://urldefense.com/v3/__https:/link.springer.com/article/10.1007/s11892-021-01375-7__;!!KGKeukY!2X8iwvcZaThROZKzTShLM48XXAAcboXdv6Nd6eTUed_wyUTOhUxUGhsoBXCoSXWbfJXR7CxWoS34HSLrfILeEy0e1iO_WRko7ct0-yQ3BcMzZ9HaWQ$>
[cid:image002.jpg at 01DB8CD6.BB1CD300]
Timothy P. Szczykutowicz, Ph.D., DABR
Associate Professor
Departments of Radiology, Medical Physics and BME
University of Wisconsin Madison
Cell# 1-716-560-7751
Office# 1-608-263-5729
he/him/his
This message contains information from Belfast Health And Social Care Trust which may be privileged and confidential. If you believe you are not the intended recipient any disclosure, distribution or use of the contents is prohibited. If you have received this message in error, please notify the sender immediately. This email has been scanned for the presence of computer viruses.
________________________________
“This message was received from outside of the organization. Please pay special attention and practice care when clicking on any links, or providing any information to the sender. Cyber attacks commonly attempt to trick you in to thinking the sender is a reputable individual who you can trust.”
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20250304/3746ca87/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image001.png
Type: image/png
Size: 3605 bytes
Desc: image001.png
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20250304/3746ca87/attachment.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image002.jpg
Type: image/jpeg
Size: 115105 bytes
Desc: image002.jpg
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20250304/3746ca87/attachment.jpg>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image003.png
Type: image/png
Size: 38606 bytes
Desc: image003.png
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20250304/3746ca87/attachment-0001.png>
More information about the Intl_dxmedphys_wd_osu_list
mailing list