Could someone eyeball this and tell me if I have missed something?
Shielding for a panoramic unit: I have the manufacturers specs which give upper range patient doses of 200 mGycm^2, and I have a 2018 paper reporting patient scatter at 1 m for 80 kVp beams as 9.3 nGy/(mGycm^2).
Assuming 50 pt / wk and rounding 9.3 to 10, I get a scatter of 10 x 200 * (100 / 2) =
1e5 nGy or 0.1 mGy per week.
And that would go to about 0.025 mGy at 2 meters. If the workload were 4/5 lower, 40 pt/wk, the unshielded scatter would be 0.02 mGy, the threshold for no lead if all fully occupied uncontrolled spaces are more than 6' away.
Its been a long time since I looked at pano shielding - any suggestions are appreciated.
---
Thanks,
Gary Isenhower