[Intl_DxMedPhys] [EXTERNAL] Re: Siemens X.Cite

Tao, Ashley T attao at gundersenhealth.org
Wed Feb 12 09:24:54 EST 2025


Hey Danielle,
Did you by chance try the using the CTDI protocol?  Most clinical protocols on the Xcite are in helical mode - Siemens has a specific CTDI protocol in axial mode.
I also noticed that this particular UI is not as easy to use as the prior UI. I found that sometimes the CTDI protocol defaults to two images (you have to manually make the slice prescription smaller so there is only 1 image). Once I manually change that, I resave the protocol so I don't have to change it every time I test.

Ashley

________________________________
From: Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list-bounces at lists.osu.edu> on behalf of Goetz, Danielle via Intl_dxmedphys_wd_osu_list <intl_dxmedphys_wd_osu_list at lists.osu.edu>
Sent: Tuesday, February 11, 2025 12:30 PM
To: intl_dxmedphys_wd_osu_list at lists.osu.edu <intl_dxmedphys_wd_osu_list at lists.osu.edu>
Subject: [Intl_DxMedPhys] [EXTERNAL] Re: Siemens X.Cite

Attention: You have received this email from someone outside of Gundersen. Please use caution responding, opening attachments or clicking on links. Report this email using the "Report Phish" button in Outlook if you are uncertain about it, if it asks for an account validation, password reset, or claims to be a security alert.







Good afternoon.



Is there a trick that anyone has found when testing beam width on the Siemens X.Cite CT scanners?



These systems are not as user friendly as others and you are limited to factor collimation settings.  The collimation 64x.6 that I tested should have given me approx. 38.4mm and I measured 61mm.  I know I need to pick a protocol without z-sharp, but has anyone else come across this and what did you find/ use for testing?





Thanks in advance.



DG



Danielle Goetz, MS

Medical Physicist

Sanford Health Bismarck

Direct: (701) 323-8722

Mobile: (701) 527-2835

Route: #20739

Danielle.M.Goetz at sanfordhealth.org<mailto:Danielle.M.Goetz at sanfordhealth.org>

[Icon  Description automatically generated with medium confidence]



-----------------------------------------------------------------------
Confidentiality Notice: This e-mail message, including any attachments,
is for the sole use of the intended recipient(s) and may contain
privileged and confidential information. Any unauthorized review, use,
disclosure or distribution is prohibited. If you are not
the intended recipient, please contact the sender by reply e-mail and destroy
all copies of the original message.
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20250212/7992d872/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image001.png
Type: image/png
Size: 5516 bytes
Desc: image001.png
URL: <https://lists.osu.edu/mailman/private/intl_dxmedphys_wd_osu_list/attachments/20250212/7992d872/attachment.png>


More information about the Intl_dxmedphys_wd_osu_list mailing list